Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 4 of 4 1 2 3 4
MikeX #73179 07/05/18 1:22 PM
Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

That's good information Mike ... thanks for posting it. smile

Are any of them battery powered? think

I was also wondering about BC Group (I like their kit); but I don't see any mention of 62353 there. frown


If you don't inspect ... don't expect.
Umi #73180 07/05/18 7:31 PM
Joined: Feb 2007
Posts: 303
Likes: 17
Master
Offline
Master
Joined: Feb 2007
Posts: 303
Likes: 17
I don’t like most of the BC Group products as they only perform the protective conductor resistance test using 10mA. The minimum for many standards, like 62353, is 200mA. I see they do one model that is a re-badged Gossen Metrawatt device, so assume this will cover 62353.

MikeX #73181 07/05/18 10:10 PM
Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

Yes; well spotted. I believe that many BC items are re-badged from OEMs. I suspect Dale in the case of the "10 mA" testers you mention. But their range overall is more or less complete, and their catalogue is very good.

I assume that the Gossen Metrawatt is the SA-2500. The manual for that one does indeed mention 62353 (and VDE 0751-1), as it does the "Alternative Measurement Method". As well as a 200 mA PE test current.

Meanwhile, here is a handy source of BC User Manuals. smile


If you don't inspect ... don't expect.
Umi #73357 09/06/18 1:32 PM
Joined: Feb 2007
Posts: 303
Likes: 17
Master
Offline
Master
Joined: Feb 2007
Posts: 303
Likes: 17
Another meter you can add to the list is the Metrel MI 3360 M. (Must have the M designation so it includes the IEC 62353 tests). Also the GMW TG euro 1 med and the Gossen Metrawatt Secutest Pro (M7050-V003) can perform IEC 62353 testing including applied parts. I see the GMC Secutest Pro is also listed on the RS website.

Also to answer the original posters question, IEC 62353 (BS EN 62353) should be used in the UK as mentioned in the MHRA Managing Medical Devices (2015) publication. This states not to use 60601-1 tests (page 22) and to use 62353 (page 24).

MikeX #73358 09/06/18 1:42 PM
Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

Originally Posted by MikeX

Must have the M designation so it includes the IEC 62353 tests


M for Medical? think

And yes; that Secutest Pro really looks "the business"! smile


If you don't inspect ... don't expect.
Page 4 of 4 1 2 3 4

Moderated by  Huw 

Link Copied to Clipboard
Who's Online Now
0 members (), 2,080 guests, and 58 robots.
Key: Admin, Global Mod, Mod
Newest Members
AndrewS1, Lam, Danny1621, zeedox, Eman Abdallah
10,386 Registered Users
Forum Statistics
Forums26
Topics11,274
Posts74,579
Members10,386
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5