Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 2 of 3 1 2 3
Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

What tolerances do those august (money-making) bodies allow, Neil? think

And what techniques do they use to weigh the weights before issuing any certificate(s)?

Also, where are test weights tested, and how are they transported etc.?

But anyway (and more importantly), what tolerances are permitted clinically for patient weighing scales?

Bearing in mind that, wherever you are in the world (and as mentioned a few time before), 1 litre of water weighs 1 kg. smile

1 litre of pure water at 4 °C weighs 1 kilogram.


If you don't inspect ... don't expect.
Joined: Feb 2009
Posts: 1,909
Likes: 18
Hero
Offline
Hero
Joined: Feb 2009
Posts: 1,909
Likes: 18
Weight & Measures, surely you have heard of them, how many people have been conned in the markets due to 'dodgy scales'

All calibration centres must be certified by the regional government offices.

Tolerances are from the manufactures specifications


I am not Flippant, I am Smart
Joined: Jun 2009
Posts: 796
Likes: 13
Philosopher
Offline
Philosopher
Joined: Jun 2009
Posts: 796
Likes: 13
We used to send weights for certification every 5 years not annually and I still think that was to much. Once they have been certified, unless you can see that a lump has been knocked off I really don't see the point in re-certification.

Any how, at 5 years our auditors seemed quite content.


Thoughts and information provided on this forum are mine and mine alone and do not necessarily reflect the policy of NSW Health. They may also be complete bollocks!!
Joined: Feb 2009
Posts: 1,909
Likes: 18
Hero
Offline
Hero
Joined: Feb 2009
Posts: 1,909
Likes: 18
Dave,
It appears that the auditors have been replaced by accreditation assessors, if they allow more time between calibrations and certification renewals they will not make as much money.


I am not Flippant, I am Smart
Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

When I asked about "what tolerances do they allow", I was referring to the "accreditation" people. frown

My suspicion is that modern patient weighing scales may be more accurate than the techniques employed by the "accreditation" labs.

In my opinion, nonsense such as this should be challenged whenever and wherever it appears. I believe it to be yet another scam based upon very little clinical need (or possibly none at all).

Newtonian physics should suffice! smile

Meanwhile, what about precision balances such as those found in clinical laboratories and pharmacies ... do the "accreditation" gang poke their noses in there as well? Or do they limit their interests to "low hanging fruit"?


If you don't inspect ... don't expect.
Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72
Originally Posted by Neil Porter

... how many people have been conned in the markets due to 'dodgy scales'


Only once ... at the souk in Jeddah. whistle


If you don't inspect ... don't expect.
Joined: Oct 2019
Posts: 2
Newbie
Offline
Newbie
Joined: Oct 2019
Posts: 2
..all very well but how to measure exactly 1 litre of water, and also what weight are the vessels holding said water?

Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

Welcome to the forum, Andre. smile

I'll leave it to you to work out solutions to those questions for yourself.

Be sure to get back to us with your suggestions.


If you don't inspect ... don't expect.
Joined: Oct 2006
Posts: 400
Likes: 11
Sage
Offline
Sage
Joined: Oct 2006
Posts: 400
Likes: 11
After repairing a set of scales that were out of warranty, and testing them with calibrated weights.

I was 'threatened' with prison by a certain scales supplier.

STILL waiting for the police to knock down my front door............

Joined: Oct 2019
Posts: 2
Newbie
Offline
Newbie
Joined: Oct 2019
Posts: 2
Thanks Geoff and glad to be here.

A large portion of my work involves weights and measures so was interested to see how you were measuring your water and weighing your vessels as was hoping to learn something, but to respond to your important question.

A clinical set of scales rated at Class III are defined as such in that the measurements taken over time can be used(in conjunction with other factors) to diagnose patients and prescribe drugs, and the tolerances we work to for this class of scale are:

Reading Discrepancy
0-500 Divisions +/- 1 Div
>500-2000 Div +/- 2 Div
>2000 Div +/- 3 Div

As a side note have attached a png showing the tolerances allowed for various classes of weights to help with your Q below...

Originally Posted by Geoff Hannis


What tolerances do those august (money-making) bodies allow, Neil?

But anyway (and more importantly), what tolerances are permitted clinically for patient weighing scales?



.[/size]

Attached Images
weights.PNG
Page 2 of 3 1 2 3

Moderated by  Huw 

Link Copied to Clipboard
Who's Online Now
2 members (daisizhou, vitapablo), 4,162 guests, and 40 robots.
Key: Admin, Global Mod, Mod
Newest Members
Czr, Jimmy Li, AndrewS1, Lam, Danny1621
10,388 Registered Users
Forum Statistics
Forums26
Topics11,274
Posts74,579
Members10,388
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5