am looking for a service manual for tyscor vs 2 suction machine by durr dental. the machine just stopped this morning. when plugged into power it just shows that there is power but no suction.
anyone with the manual can share with me to my vgmakwinja@gmail.com
Thanks for coming through so quickly. There are two of them. So, with the one when it is switched on it just starts to suction and after some seconds the LEDs come to RED : Fault suction and fault SEPA. For the other one the moment you switch it on the suction process does not even start it just comes straight to RED: Fault suction and Fault SEPA.
Any clues as to what could be wrong? Incase you encountered such a problem before.
Well; looking at the manual is seems that the red LED indicates a problem with either the radial blower or the separation system.
Just a guess ... but I would look closely at the separation system in the first case, and the blower in the second. Either way I would disconnect the separation system(s) and try again.
But two machines? From the same user? What story is the user giving?
I'm putting my money on flooded ... blocked ... dirty ... separation units (but hopefully not as far as the blower) - most likely due to "operator error" or lack of cleaning etc. Let's see if I'm right.
I wonder what cleansing agent(s) are being used. I should imagine that Dürr are pretty fussy in their requirements.