Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Joined: Jun 2025
Posts: 4
Newbie
OP Offline
Newbie
Joined: Jun 2025
Posts: 4
We've been informed by Cosmed that they no longer supply replacement cell kits for the K4b2.

Unfortunately, we can't justify spend on the K5 as the K4 is only used occasionally for teaching and student projects.

Looking at the O2 cell, it's looks like an off-the-shelf cell that's been slightly modified to have right angled ports instead of the threaded connector. Unsure of what cell it could be however.

[Linked Image from i.imgur.com]
[Linked Image from i.imgur.com]
[Linked Image from i.imgur.com]

Joined: Oct 2011
Posts: 279
Likes: 22
Master
Online Content
Master
Joined: Oct 2011
Posts: 279
Likes: 22
If you can measure the mv voltage curve corresponding to different oxygen concentrations?
You can use compatible oxygen sensors or modify the proportional resistance value inside the oxygen cell to meet your requirements.


daisizhou#sina.com (#=@)
Joined: Sep 2004
Posts: 138
Likes: 3
Expert
Offline
Expert
Joined: Sep 2004
Posts: 138
Likes: 3
Try Steve (Technical Support Manager) at Viamed - he was really helpful when I was looking for an oddball fuel cell.



I laugh in the face of danger. Then I hide 'til it goes away.
Joined: Jun 2025
Posts: 4
Newbie
OP Offline
Newbie
Joined: Jun 2025
Posts: 4
Originally Posted by daisizhou
If you can measure the mv voltage curve corresponding to different oxygen concentrations?
You can use compatible oxygen sensors or modify the proportional resistance value inside the oxygen cell to meet your requirements.


I don't think that would be reliable, as far as I'm aware these cells begin to degrade once breaking the seal. They are marked with an expiry date. We don't have a new one to gather reliable values from, unfortunately.

Joined: Jun 2025
Posts: 4
Newbie
OP Offline
Newbie
Joined: Jun 2025
Posts: 4
Is there a direct address to get him on? Or should I just contact their technical address?

Joined: Sep 2004
Posts: 138
Likes: 3
Expert
Offline
Expert
Joined: Sep 2004
Posts: 138
Likes: 3
DM sent



I laugh in the face of danger. Then I hide 'til it goes away.
Joined: Oct 2011
Posts: 279
Likes: 22
Master
Online Content
Master
Joined: Oct 2011
Posts: 279
Likes: 22
Originally Posted by Chris DArcy
Originally Posted by daisizhou
If you can measure the mv voltage curve corresponding to different oxygen concentrations?
You can use compatible oxygen sensors or modify the proportional resistance value inside the oxygen cell to meet your requirements.


I don't think that would be reliable, as far as I'm aware these cells begin to degrade once breaking the seal. They are marked with an expiry date. We don't have a new one to gather reliable values from, unfortunately.
You need a millivolt generator,Connect the millivolt voltage generator to the original oxygen battery connection port.Observe the oxygen concentration displayed by the host and record the corresponding millivolt voltage value at this time.

Complete the curve from 0%-21% oxygen concentration to 100% oxygen concentration.

Then buy other types of oxygen cells to match the curve you measured and you are done.


daisizhou#sina.com (#=@)
Joined: Jun 2025
Posts: 4
Newbie
OP Offline
Newbie
Joined: Jun 2025
Posts: 4
Originally Posted by Rob1234
Try Steve (Technical Support Manager) at Viamed - he was really helpful when I was looking for an oddball fuel cell.

Sadly, Steve left just last week according to Viamed. No one else would help because it's a "modified special". I suggested I could try to give them information to inform on a replacement (that we could modify in house) but no such luck.


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 3,379 guests, and 14 robots.
Key: Admin, Global Mod, Mod
Newest Members
Szymon1990, Janet, james_on_ebme, Biomed Egypt, Nektarios K.
10,365 Registered Users
Forum Statistics
Forums26
Topics11,255
Posts74,507
Members10,365
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5