Hi Colleagues,
Infusion pump measurement is a very interesting discussion.
At North Middlesex Hospital we've used the two manual methods to test our Arcomedical Volumed uVP5005 pumps i.e the measuring cylinder (burrett? - I can't spell either!) and also very accurate weighing scales (e-Accura SC131), always using distilled water as the pumping medium to measure the total volume delivered. In addition, the total time of delivery is measured using an electronic bench stopwatch so that the rate of delivery can be checked. The method is to connect the stopwatch to the "nurse call" socket on the pump. Start the pump and the stopwatch together and when the pump finishes it's delivery it automatically stops the stopwatch. Hence, the total time that it takes to deliver the set volume can be read (without sitting watching the pump with a manual stopwatch!) Therefore, the average rate can be checked as total volume/time (ml/h etc)
Just reading the volume in the measuring cylinder isn't very accurate at all. Our 100 ml measuring cylinders have only 1 ml graduations and are only accurate to +/- 0.5 ml @ 20 deg C. Moreover, it is very difficult to read the level accurately from the cylinder by eye. Using the cylinder with the accuate scales is much better. Moreover you can "zero out" (tare) the weight of the cylinder when it's empty and hence get a direct reading from the scales of the weight of the volume of dist. H2O delivered.(This method is clearly a cheaper option than purchase of an IDA device.)
What I'm not clear about is, do we need to measure the instaneous rate as opposed to the average rate delivered as you would with an IDA-2 or IDA-4 device? The use of of this type of testing device is obviously much better, especially if your testing lots of pumps.
However, it seems to me that with all testing/checking we should only carry tests that the manufacture recommends to ensure that the equipment is working within its specification. Volumed do not specify that to test their pumps you must use an IDA device. In fact their engineer has only one rate check on his test sheet i.e. 100ml/h with VTBI = 50 ml. We check this with +/- 1% accuracy (49.50 - 50.50 ml with an expected period of 29.40 - 30.20 min). Possibly, too accurate a measurement?
Finally, it's obviously important to always use a new or hardly used giving set when testing pumps.
Having said all of the above I'm sure that our unit would benefit from having a new IDA device given the number of infusion pumps that we test. (We've got a very old IDA-2 sitting around gathering dust. I think I've seen it used more than a couple of times about 10 years ago as it always had problems!)
Ian Rez