2 + 1 = 3 (terminals - one +ve, one -ve, one grounded)
The third one is not ground it is usually active (Right Leg Drive) to get rid of common mode noise. The actual signal is the difference of the other two electrodes.
If you look at Holter monitors they usually have two pairs of two leads, each pair giving a signal. No third electrode needed in this fully floating environment.
Robert