Joe,
The unit was eventually checked without any BNC’s connected. Still the fault was apparent.
Mark,
You are correct; these sockets should each individually be clearly identified.
WOM, Ken, Nick, Ged, Grahame and Roy
The eventual resolution was for the company have offered two solutions to the problem.
1) To replace the 14-inch monitor with a 21 inch one. Is seems the 21 inch monitor did not have this problem.
2) To fit an isolation transformer to the trolley, therefore being able to use the current 14-inch monitor.
During a visit by the Starkstrom engineer several weeks ago (Who I must add was extremely helpful) came up with a probable theory.
If you recall my initial posting
“The device installed at the intake (Consumer unit) is an IPSM002-MB.
What this device does is monitor the insulation resistance of the unearthed AC network, mainly found in group 2 medical locations.
The current temperature of the isolation transformer is permanently measured. It is in the form of a DIN rail mounted device.
Its function is to measure DC voltage that is superimposed on the network”
His theory is that this particular model of Sony monitor emits a DC voltage back onto the line. This in itself interferes with the Isolated mains monitor system (IPSM002-MB) installed by his company?
This theory seems to ring true, and we proved it by using an isolation transformer fed from the theatres supply with the 14inch Sony monitor without any hindrance.
However this kind of defeats the principle of having this ITPS monitor (IPSM002-MB) power system installed in the first place eh?
Using an isolation transformer to isolate it from the Isolated (IPSM002-MB)Power system ?
Chris-H
P.S I did post an earlier reply on this subject, however I believe it was at around the time the server was backed up by HUW following the crash.