Hi All
we are having problems with the cleaning of a GE TV probe where the insulation is going yellow and cracking at the strain relief end after about 6 months. The probes are wiped down with clinell wipes dried and then put in a Trophon unit. The Clinell wipes and the Trophon Units are authorised products by GE for cleaning their equipment. I have looked at other probes in a different department and their GE probe cables are showing sign's of yellowing which leads me to the Clinell wipes. GE are blaming the cleaning method for the problem which I disagree with. Our TV probe is used daily and is cleaned about 15 times a day.

Does anybody use Clinell wipes and Trophon units together?
If yes do you have any issues?
Do you use either of the above?
if yes are there any issues?

regards

richard