Hi Tommy, please remember to test the UFP (Bovie) outputs and also output 2, they are not specified to be tested in the manual, but the users may well be using them and they need testing.
One common issue i see is the users plug the footswitch into socket 1, and then the patient cable to socket 2. Footswitch socket 1 is for output 1, socket 2 is for output 2.
The PM is fairly easy to perform taking around 30-40 minutes.