Good Morning All,
I hope you are well. Is anyone else experiencing issues when testing the above.
"There has been an error that the devices are not picking up CPR on the final test on the test reports. Zoll have been contacted and they have said a common cause of this is the pins on the one-step cable being bent/broken and shorting out the CPR board inside the device. Error code say(I2C OP ERR) which indicates a CPR board failure.
The error presents itself about 10 seconds after the device is turned on with pads or the leads that connect to the defib tester.
You will have PERFORM CPR - CHECK CPR PUCK - I2C OR ERR - IF NO PULSE come on to the screen and the CPR release + PPI will not react to any pressure from the transducer when applied for the test. " Zoll are saying we cannot replace this board ourselves and are charging us an extortionate 3,300 per device for this. Massively grateful for any info anyone can offer us . Have a fab weekend