Planned maintenance could be categorised in the following 3 areas: 1. High risk device 2. Medium risk devices 3. Low risk devices
Risk categorisation could be determined by:  looking at clinical usage  reliability  dangers to patient on failure  User training
A Risk centred common sense approach allows better use of resources.
High risk These device are life saving / threatening. [e.g. Defibrillators, Infusion pumps] All high risk devices will be checked regularly for safety and funtionality, technicians will receive technical training of a suitable standard.
Medium risk These devices aid diagnosis and therapy [e.g. ECG, Physiotherapy] Some medium risk devices may be maintained
Low risk These devices aid diagnosis [e.g. Handheld Aurascope, Aneroid sphygmomanometer]
PPMs on high risk devices should be carried out more frequently than PPMs on Medium risk devices. PPMs on Low risk devices may not carried out if the risk is deemed to be acceptably low.
We are already using this system to manage our ppm list. What is the concensus? (I know the MDA ONLY recommend following manufacturers advice) How do you manage your ppm's?
If PPMs cannot be carried out for any reason, should the EBME Manager use his experience and professionalism in deciding which PPMs are considered necessary?
Be Proactive and reactive.
|
|
|
|
|
Joined: Jul 2001
Posts: 91
Adept
|
Adept
Joined: Jul 2001
Posts: 91 |
John, I have spent considerable time on this issue and have come up with a system which builds on the empirical work already in use in the UK and US. This has been ratified by my Risk committee as an appropriate means of quantifying risk, and optimising resource use. It has been peer reviewed by the NPAG southern benchmarking (EBME)club and has been proposed as 'best practice'
It basically derives a risk score in a similar way to you (High, Med,Low)which is used to prioritise/guide resource utilisation Planned maintenance could be categorised in the
It considers several risk factors;
Functional, Physical, Location, and overlays them with maintenance, utilisation, and mission scores (mission = degree of importance to Trust in providing a clinical service)
Obviously there is a lot more to it than assiging scores/weightings. The important thing is that is applies consistency of approach, optimises management of (device)risk, and provides a ready set of performance indicators, as well as the data needed to justify modification of some criteria (maintenance frequencies, relative change in importance to Trust in meeting clinical outputs).
'If PPMs cannot be carried out for any reason, should the EBME Manager use his experience and professionalism in deciding which PPMs are considered necessary?'
Yes, but it needs qualification in light of the risk factors outlined + justified from such evidence. Whilst extensive experience + gut feel are often good enough, trying to justify that to 'the one that got away' to a court will not cut any ice.
Morale = document, document, document.
If you want further info, I am happy to provide
Paul H
|
|
|
|
|
Joined: Jul 2002
Posts: 499
Sage
|
Sage
Joined: Jul 2002
Posts: 499 |
John, Paul, Louis employs a similar algorithm to the one's described. Louis's hospital FORMALLY classifies/grades equipment in a similar countenance which John describes.  This classification is, Louis believes, an American system and it is quite often written in to the relevant specification i.e. (as Johns, and I believe he has worked where Louis is now) Class 1. High risk device 2. Medium risk devices 3. Low risk devices Louis's can then prioritize accordingly though Louis still likes to fulfill his PPM agenda as do we all, my friends, as do we all.  If you can get your property managment etc. to formally grade equipment, you may find it easier to prioritize on a more qualitive podium than that of your personal “gut instinct”. Louis III 
No trees were harmed in the posting of this message. However, a large number of electrons were terribly inconvenienced. كيف الآن يحمّر البقرة
|
|
|
|
|
Joined: Jul 2000
Posts: 1,969 Likes: 32
Hero
|
OP
Hero
Joined: Jul 2000
Posts: 1,969 Likes: 32 |
Paul, Louis, I'm pleased to see that we are thinking along the same lines. I would like to carry out this function in the most professional way, and I would like to see your 'ratified' risk process, which is leap-years ahead of mine. Paul, Louis, if it is in a format which can be e-mailed could please you send me a copy to: john@ebme.co.uk , also would you like to put them up to share as downloads? If so, let me know when you send them. 
Be Proactive and reactive.
|
|
|
|
|
Joined: Jul 2001
Posts: 91
Adept
|
Adept
Joined: Jul 2001
Posts: 91 |
John, I have mailed you a couple of items which will help explain. I think that the presentation will need significant modification to make it self explanatory to a wider audience. Please give your views before consideration is given to doing this PAul
|
|
|
|
|
Joined: Jul 2000
Posts: 1,969 Likes: 32
Hero
|
OP
Hero
Joined: Jul 2000
Posts: 1,969 Likes: 32 |
Paul Thank for the info. I will have a read and get back to you.
Be Proactive and reactive.
|
|
|
|
|
Joined: Jul 2002
Posts: 499
Sage
|
Sage
Joined: Jul 2002
Posts: 499 |
John, its a public holiday where Louis dwells,  Louis will e-mail you our full "equipment classification policy" when Louis's Gluteus is firmly back at his desk. Meantime I believe it was tailoured around the The C.E. Medical Devices Directive guidelines:- Classification
The directive applies different administrative requirements to devices depending on the risk which they present to the user. There are four classifications:
Class I - generally regarded as low risk Class IIa - generally regarded as medium risk Class IIb - generally regarded as medium risk Class III - generally regarded as high risk.
Once it has been decided that a device falls within the scope of the Medical Devices Directive, the actual classification of an appliance is determined by a series of rules which are laid out in annex IV to the Directive. Essentially these start by saying that anything which is not covered by one of the other rules is a class I device, and then go on to give more detailed rules for the classification of other equipment. Typical factors upon which the rules decide are whether the device is intended for long term or short term use, whether it involves the use of associated pharmaceuticals and chemicals, whether it is invasive in normal bodily orifices or via surgical insertion, and whether it operates by the exchange of energy with the body. Definitions of terms used to describe these factors are also included in the annex. You could contact cemark@conformance.co.uk for an extensive listing or use the following:- CE Conformance. Great Hucklow Buxton Derbyshire SK17 8RG England Tel: +44 (0)1298 873800 It works really well for Louis's institution and combined together with a risk assesment policy it is the "dogs b*lloc#s" in terms of "labour shortage" with regards to gluteus habiliment Kind Regards Louis III 
No trees were harmed in the posting of this message. However, a large number of electrons were terribly inconvenienced. كيف الآن يحمّر البقرة
|
|
|
|
|
Joined: Jul 2001
Posts: 91
Adept
|
Adept
Joined: Jul 2001
Posts: 91 |
Louis, I follow the device classification logic, + therefore the resource utilisation argument. But how do you factor in the maintenance and environment (things important/unique to your organisation) variables + how do you review/justify changes to PV/PM frequencies etc
I am observing + desperately hoping to learn !!
Paul
|
|
|
|
|
Joined: Jul 2002
Posts: 499
Sage
|
Sage
Joined: Jul 2002
Posts: 499 |
Paul, my coadjutor, if by environment you symbolize “location” then yes that has always and will continue to be a drawback,  however, every clinical department in Louis's trust sits an entrusted person  who is responsible for ensuring that items are available as and when they are due.(thanks KM)  , Louis appreciates that a lot of trusts probably don't employ this type of system but it really does work. There are also added benefits of EBME-USER relations which Louis feels are strengthened by this arrangement. Our computer system is managed by our maintenance control room, this includes all PPM's and general maintenance, they distribute the work to their relevant departments i.e. electrical, mechanical, EBME etc. (Note EBME is totally independent of all other services) (Note EBME is totally independent of all other services) It is an excellent system but the initial cost is mega wonga,  but it allows us to reschedule work if necessary. As there are in comparison only a small number of “in-accessible” pieces of equipment it hardly effects the annual distribution of work. Hope this is informative my friend Louis III
No trees were harmed in the posting of this message. However, a large number of electrons were terribly inconvenienced. كيف الآن يحمّر البقرة
|
|
|
|
|
Joined: Jul 2000
Posts: 1,969 Likes: 32
Hero
|
OP
Hero
Joined: Jul 2000
Posts: 1,969 Likes: 32 |
Louis, Paul, Do you ever have problems trying to convince staff that their equipment no longer needs a PM check? (When classified as LOW RISK) I currently have someone here trying to insist on their equipment having PM's.
Be Proactive and reactive.
|
|
|
|