Thanks Chris - I wasn't aware of that second news story. My questions above have been answered.
From my experience with the Baxter Colleague the main problems are those associated with batteries that fail in use - they split, overheat and sometimes, in fact regularly, leak into the pump (see previous links from way back, elsewhere in the forum). The LA battery used in these pumps is notoriously unreliable in my experience - it also failed regularly in a similar fashion in another company's pulse oximeter products I used to work on. I don't know whether poor design is a contributory factor or not but it's possible.
Errors that I assume are associated with power failure during use, either on battery or on mains are apparent and a regular occurrence. On older models with early software this can result in the sets being trapped in devices even when the pump is still functioning when the giving-set mechanism resets.
Under these circumstances, when devices fail, operators panic, can't remove giving sets that are connected to patients, can't turn devices off (if they're alarming) unless they manually reset the mechanism (which is not necessarily obvious if there are no prompts or guidance available from the pump) or turn them back on again (if they power-down with the mechanism closed) with the set still within the pump. That's why we get cases where sets have been cut and fluids get inside pumps when they fail.
The early-revisions of battery harnesses allow currents to circulate between LA batteries that may have different levels of charge (due to mismatch in capacity and different battery terminal voltages I assume) - thus, in dual battery models, faulty batteries can deplete or compromise batteries that are Ok. Potentially high currents (up to the 4A inline fuse limit) can circulate in paralelled batteries until equilibrium is reached, which is poor design, in my opinion.
I'm aware of the availability of replacement battery looms with current-limiting and reverse-protection devices fitted in them and the availability of replacement software to overcome problems with operators handling pump mechanism failures in use. As far as I'm aware Baxter UK are making these looms and software available and have circulated information concerning battery maintenance issues.
This is a device that some "experts" in procurement who work in NHS supplies, EBME and Medical Physics departments have been touting for quite a while I think. Someone hasn't been doing their homework and keeping up with developments, perhaps. I think it's our duty to make our operators aware of the potential issues if the MHRA, risk managers and supplies departments do not.
The Baxter recall, as communicated by the MHRA, doesn't make reference to the battery and pump head mechanism related failures (which I and others I've spoken to consider to be the common causes of failure in use) does it?