Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 2 1 2
Joined: Jan 2002
Posts: 21
JohnA Offline OP
Dreamer
OP Offline
Dreamer
Joined: Jan 2002
Posts: 21
Hi,
We've had several instances recently where patients/relatives have been discharged with medical equipment from our loan library e.g syringe driver or volumetric pump.
Obviously they had training on the device before being allowed to take it. I am aware of the MHRA "checklist for patients discharged from hospital with a medical device"
What do other Trusts do in cases such as this? Do you make them sit some sort of competency test? How do you know that in the case of a adverse incident you're not in the poo?
Thanks, John

Joined: Jul 2000
Posts: 1,968
Likes: 32
Hero
Offline
Hero
Joined: Jul 2000
Posts: 1,968
Likes: 32
Good question. We leave it to our nurse specialists who look after the patients. Hopefully they do it properly. I will check.


Be Proactive and reactive.
Joined: Oct 2003
Posts: 26
Dreamer
Offline
Dreamer
Joined: Oct 2003
Posts: 26
How do you manage with the problem of servicing the equipment whilst it is in community? We look after a number of community patients with a wide variety of equipment and are at present chasing up service records.

Joined: Feb 2002
Posts: 43
Technologist
Offline
Technologist
Joined: Feb 2002
Posts: 43
E-MAT can be interfaced to any Patient Administration System. At our Dewsbury site when a piece of equipment is due to be serviced they can see the demographics of the patient via the Equipment Loans module for address details etc.

Regards,

Paul Stockton
Barwick Systems
http://www.barwicksystems.co.uk
ISO 9001 Registered Firm

Joined: Mar 2004
Posts: 28
Sur Offline
Visionary
Offline
Visionary
Joined: Mar 2004
Posts: 28
Hi JohnA,
Your question about how the patients/carers are trained or their competency assessed, is a very genuine question and people seem to shy away from it. I have been looking into this area for some time and found that the practices vary across the NHS. What I can say for sure is that trusts do train patients/carers before a medical device is issued to them, however, I have not come across any example where the patients'/carers' competency have been assessed. Again, regarding training, I don't think there is any 'systematic' approach. People have to resort to 'common sense' method such as demonstration. And I am also not aware of any standards or guidance either from the DH or from the MHRA which depict how the patients/carers should be trained or their competency assessed. I hope other experienced people out there may put more light on this matter.

Joined: Aug 2004
Posts: 6
Newbie
Offline
Newbie
Joined: Aug 2004
Posts: 6
Hi JohnA

Within our Trust I have just been delegated the task of assessing and ensuring that "all end users (patients) have access to manufacturers instructions and that users sign statements to the effect that they have recieved instructions on the safe use of devices or equipment." Where carers are taking home infusion devices they will attend a training workshop (group or individual) and they will have their competence assessed. I will then keep a record of this training.

So I guess Im about at the same stage as you with this one!


Sharan
Joined: Mar 2004
Posts: 28
Sur Offline
Visionary
Offline
Visionary
Joined: Mar 2004
Posts: 28
Hi Sharan

Could you please let us know about what is imperative (legally obligatory) and what is just a 'good-practice'? As you said "all end users (patients) have access to manufacturers instructions and that users sign statements to the effect that they have recieved instructions on the safe use of devices or equipment" Is this required by the law or is just your trust's policy? Again, what do you mean by "... they have recieved instructions on the safe use..."? Do instructions mean manufacturer's instruction materials or training? Thanks

Joined: Aug 2000
Posts: 156
Mentor
Offline
Mentor
Joined: Aug 2000
Posts: 156
Sur,
I would be interested in knowing how you conducted your research and evidence that this does not happen.

You do not need competency training on all equipment. The type of equipment in question is; MS26 for palliative care, Occassional Syringe pump and pressure mattresses. You will find all the carers and patients have been trained and certainly via EMAT we trace this to the patient detailos and PVI system for recall purposes.

the dietetic pumps fall under same category with swap out.

Our system works extremely well and nothing gets lost in the system.

I think the issue is, Do all MTOs get involved in the clinical care of the patients as part of the loop or not? At Mid Yorks, all Medical Physics Staff form part of the chain, hence the reason for having a robust system.

Our Librarians, visit the wards on daily bais and talk to staff (across all 4 sites), big job, but we do it and it works and all info is inout into EMAT library module.

hope this helps. The message is, get out there, work will all clinical staff including consultants and up to Medical Director Level.

alex

Joined: Aug 2000
Posts: 156
Mentor
Offline
Mentor
Joined: Aug 2000
Posts: 156
sorry on the paragraph before last, inout should read input.

alex

Joined: Feb 2004
Posts: 14,813
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,813
Likes: 72
Why not use the edit facility, Alex?

It's the paper and pencil symbol next to the quotes " " (and I've just used it to edit this).

See https://www.ebme.co.uk/forums/ubbthreads.php/ubb/faq
smile

Quote:
Users can edit their own posts up to 120 minutes after they are made. Where the change is substantial, it is courteous mark the post as edited so as to alert readers to the changed content. For cosmetic changes, it's better not to do so.


If you don't inspect ... don't expect.
Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 3,201 guests, and 12 robots.
Key: Admin, Global Mod, Mod
Newest Members
Danny1621, zeedox, Eman Abdallah, OmarLY, Hans Pauwels
10,384 Registered Users
Forum Statistics
Forums26
Topics11,274
Posts74,576
Members10,384
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5