Dear All
We are looking at ultrasound units for vasular access re. the NICE guidance. I would welcome any feedback on the makes/models are you using.
Thanks
Bob..

Joined: Feb 2003
Posts: 380
Sage
Offline
Sage
Joined: Feb 2003
Posts: 380
We use the Sonosite iLook 25. PaSA has a model configured for this particular application on its framework agreement. Last time we bought one it cost about £9K

Also it's a generic portable U/S unit so can be used for other applications if you have the right probe.

Joined: May 2005
Posts: 15
Novice
Offline
Novice
Joined: May 2005
Posts: 15
dear bioman,
this sonolite i look ,which manufacturer,any other manufacturer ? anyone knows ,please inform.

ashoke

Joined: Feb 2003
Posts: 380
Sage
Offline
Sage
Joined: Feb 2003
Posts: 380
Sonosite is the manufacturer and iLook 25 is the model. I think they are an American company.

Any generic ultrasound can be used for catheter placement as long as you have an appropriate probe. Therefore if you have a machine already you just need the right probe, which you may already have so no additional cost.

The iLook is ideal for dep/wards who want to use this technique but don't have an existing U/S machine as it is reasonable priced for a ultrasound. There are other cheap options but they have screens which are small and poor quality.

Joined: Jul 2000
Posts: 969
Likes: 1
Roy Offline
Philosopher
Offline
Philosopher
Joined: Jul 2000
Posts: 969
Likes: 1
Sonosite Ltd., Alexander House, 40A Wilbury Way, Hitchin, Herts. SG4 0AP
Tel: 44 (0)1462 444801
Fax: 44 (0)1462 444801

Very helpful company - good equipment.


Today is the day you worried about yesterday - and all is well !
Joined: May 2005
Posts: 7
GK Offline
Newbie
Offline
Newbie
Joined: May 2005
Posts: 7
How about the vascular doppler units..?
For example from Nicolet vascular?

These will be much economical than the iLook full ultrasound units.
The only difference, these are non imaging...

A
Anonymous
Unregistered
Anonymous
Unregistered
A
The whole point is the imaging aspect I think.

Joined: Feb 2003
Posts: 380
Sage
Offline
Sage
Joined: Feb 2003
Posts: 380
Some vascular doppler only have sound so would not be good for catheter placement. The idea of using U/S for catheter placement is so that you can see that it has gone down the right way and hence why NICE are advising this technique as opposed to the traditional landmark technique i.e. lets insert for about 10cm and turn right, doing the landmark technique runs the risk of the catheter ending up in the wrong place.

Joined: Dec 2001
Posts: 38
Visionary
Offline
Visionary
Joined: Dec 2001
Posts: 38
We have used the Siterite from Jade medical for many years with very few problems. Its a nice old fashioned mechanical sector scanner and gives a very clear image for catheter placement.


To err is human; to do so more than once is contrary to departmental policy.
Joined: Mar 2008
Posts: 16
Novice
Offline
Novice
Joined: Mar 2008
Posts: 16
You could also try the Mindray DP-6600 with a 5-10MHz linear array probe. Simple to use and inexpensive at around £5k + VAT

You can contact me at www.caiysideimaging.co.uk for more information if you want.

Last edited by Sonix; 24/05/08 1:03 AM. Reason: More information
Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 1,411 guests, and 28 robots.
Key: Admin, Global Mod, Mod
Newest Members
Mark B, Royal, davidarhui, Czr, Jimmy Li
10,391 Registered Users
Forum Statistics
Forums26
Topics11,278
Posts74,590
Members10,391
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5