We have a patient that they are looking to get home but they are currently requiring 10lpm of oxygen. Cylinders would be too bulky or would need replacing too frequently to be a viable option as an oxygen supply.
Does anyone have knowledge of high flow concentrators, I have only found ones that give 6lpm and then only at 80%, or liquid oxygen systems? I know of the Tyco Helios systems but have no specification of the system.
Thanks all
Robert


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
Joined: Feb 2003
Posts: 380
Sage
Offline
Sage
Joined: Feb 2003
Posts: 380
Would BOC be able to suggest a solution?

Joined: Jul 2002
Posts: 2,020
RoJo Offline OP
Hero
OP Offline
Hero
Joined: Jul 2002
Posts: 2,020
Yes, lots of cylinders, which has storage and handling problems.
You would need to change an F cylinger every two and a quarter hours or a J cylider twice a day but these weigh 70kg each.
And what would the fire brigade have to say about this amount of oxygen? They were concerned about my small oxy-acetelene setup.
Hence looking for more radical solutions.
Robert


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
Joined: Sep 2005
Posts: 5
Sky Offline
Newbie
Offline
Newbie
Joined: Sep 2005
Posts: 5
You can couple concentrators together using a T-piece to achieve the aggregate flowrate you desire

Joined: Jul 2002
Posts: 2,020
RoJo Offline OP
Hero
OP Offline
Hero
Joined: Jul 2002
Posts: 2,020
That is one solution we are looking at, Sky, but this doubles the maintenance, noise and amount of floor space needed.
Robert


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
A
Anonymous
Unregistered
Anonymous
Unregistered
A
I was considering something like this for a poorly relative with severe respiratory problems a while back (to be prescribed by a physician of course). I too did not fancy having (4 AF) full cylinders lying around the house continuously with a smoker:

BarO2 Concentrators

Still a fire risk but possibly more acceptable to the local Fire Brigade. It says up to 10 lpm. Concerns are if it is Medical Grade or not and whether output pressures are appropriate.

Also:

PaSA Domicillary O2 Risk Assessement

Joined: Jul 2002
Posts: 2,020
RoJo Offline OP
Hero
OP Offline
Hero
Joined: Jul 2002
Posts: 2,020
Richard,
I came across these people with a web search and ignored them. I have had another look and see they do a concentrator that claims to do 10lpm but I have also seen that it produces 75-99% oxygen. The medical ones start to drop off the concentration at over 6lpm so they limit the flow. I wonder if this one is similar but without the limit.
I will investigate it further.
Thank you
Robert


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
#21706 29/09/05 10:29 AM
Joined: Dec 2003
Posts: 282
Master
Offline
Master
Joined: Dec 2003
Posts: 282
Oxygen is a drug and therefore needs to be prescribed. I think there is still Pharmaceutacil legislation that limits the flow for domicilliary oxygen to a maximum of 4l/min. Hence most concentrators have a maximum useful flow of 4 l/min and domicilliary regulators are medium (2l/min) and high (4l/min.
Our Respiratory Consultant will not prescible concentrators to people who are still smoking.


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 4,664 guests, and 13 robots.
Key: Admin, Global Mod, Mod
Newest Members
Royal, davidarhui, Czr, Jimmy Li, AndrewS1
10,390 Registered Users
Forum Statistics
Forums26
Topics11,278
Posts74,588
Members10,390
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5