ROB1234,
If you are using Asena CC's, why didn't you put the pressure sensing disc in the CC set after the anti-syphon valve? Surely this would have eliminated the problem of requiring a higher occlusion limit as the sensing is done in-line?
I have had this problem with Asena GS's and Graseby 3300's.
One problem is that it seems to rear its head and then disappear for a while before resurfacing again!?
I agree with Paul that increasing the occlusion level is a solution, but if a line is used without an anti-syphon valve, the occlusion level will be too high.
This wouldn't be an issue with the old P3000's as the user could increase the occlusion alarm when required (although I have reservations as to whether the user being able to increase the occlusion alarm is a good thing!)
It appears that it is a case of the Devil or the Deep Blue Sea:
You can have an occlusion alarm to tell you when there is an interuption in delivery or you can have an anti-syphon valve to eliminate over-delivery, but using one negates the benefit of the other.
Thanks for your replies, though. At least I'm not alone in having this problem.