Hi folks
I'm after some information regarding the testing of nebulisers. As i haven't got access to any devices or manuals i'm after those of you folks who have the knowledge and what sort of pressure and flow testing you carry out.
Also what test equipment do you use for the servicing of these items and are there any specifics that might be useful. Im looking at the Medix range like the AC4000.
Any information what be of great help folks.
|
|
|
|
|
Joined: Feb 2004
Posts: 14,815 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,815 Likes: 72 |
Yes Darran, output pressure and flow are the final tests to be carried out once you have cleaned out all the crud, changed the filter, confirmed electrical safety etc. There used to be a well-known test rig (consisting of a pressure gauge, and flowmeter, mounted on a board) that I've come across in many biomed shops. I think they were supplied by Medic-Aid (in those days), or maybe it was Medix (someone out there will know)? Once you have sorted out your test rig (and you could always make your own), you will need a little chart listing the go/no-go limits for any and all nebs you service. Again, perhaps one of the Brethren may be kind enough to offer such a list right here! 
Last edited by Geoff Hannis; 13/06/07 11:35 AM. Reason: It needed editing.
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Sep 2005
Posts: 70 Likes: 9
Scholar
|
Scholar
Joined: Sep 2005
Posts: 70 Likes: 9 |
Considering you can pick up a nebuliser for £60 I would be inclined to do minimal work on these units. All nebulisers I have are checked at a pressure of 20PSI. If the flow corresponds to what the manufactures state I would check the plug, fuse, mains cable and ensure it is not physically damged. If the unit is more than 3 years old I would scrap it immediately if it has a low output, is physically damaged or even if it is particulary dirty. I wouldn’t consider going inside a particularly dirty unit and cleaning it, as I wouldn’t consider that economical.
Last edited by kit; 13/06/07 1:18 PM. Reason: Grammar
|
|
|
|
|
Joined: Jan 2006
Posts: 141
Expert
|
Expert
Joined: Jan 2006
Posts: 141 |
AC4000 nebulisers specified dynamic flow rate is at least 7 lpm @ 15 psi.
Medic Aid ( now Respironics ) supplies the testing rig
|
|
|
|
|
Joined: Jul 2000
Posts: 1,969 Likes: 32
Hero
|
Hero
Joined: Jul 2000
Posts: 1,969 Likes: 32 |
Takes 1 minute to check on a test rig. Some repairs are still worth doing, but I agree it should be limited to no more than 10 - 15 mins or throw it in the bin.We have an assitant technician that does nebs (changes filters, cracked cases, etc).
Be Proactive and reactive.
|
|
|
|
|
Joined: May 2007
Posts: 8
Newbie
|
OP
Newbie
Joined: May 2007
Posts: 8 |
Thanks for the response guys i wasn't aware of the that you could actually buy a test rig. I've not been able to find this on the web with the info you guys gave so if anyone could supply a link that would be great. Also as anyone got a list they use of the limits for nebs when testing. Anyone got a contact for things like filters and such like for Medix? as i've had no luck 
|
|
|
|
|
Joined: Aug 2001
Posts: 797 Likes: 1
Philosopher
|
Philosopher
Joined: Aug 2001
Posts: 797 Likes: 1 |
DARRAN MEDIX IS CLEMENT CLARKE, ASK FOR THEIR REPIRATORY PRODUCTS TEAM FOR HOSPITALS (ITS A DIFFERENT TEAM TO JOE PUBLIC). MEDIC AID IS RESPIRONICS ASK FOR TECHNICAL SUPPORT DEPARTMENT. IN BOTH CASES THEY ARE USUALLY VERY APPROACHABLE AND CAN USUALLY SORT ANY PROBLEMS. WE HAVE 2 TEST RIGS THAT WE BOUGHT IN 1995 FROM THE ABOVE AT £135.00 EACH. THEY ARE BASICALLY A FLOMETER AND GAUGE MOUNTED ON THE PIECE OF BOARD. YOU CONNECT THE NEB OUTLET TO A PIECE OF TUBE AND GET A READING ON BOTH.
|
|
|
|
|
Joined: Jan 2006
Posts: 141
Expert
|
Expert
Joined: Jan 2006
Posts: 141 |
KM - Don't shout here, we ain't deaf on this forum!  Darran - check your PMs
Last edited by Jonathan Wells; 14/06/07 8:14 AM.
|
|
|
|
|
Joined: May 2002
Posts: 137
Expert
|
Expert
Joined: May 2002
Posts: 137 |
Whilst the pressure guage is a standard item, the flow meter needs a calibrated nozzle in series with it as the flow (current) and pressure (voltage) specifications are stated with the correct flow restriction (resistance).
|
|
|
|
|
Joined: Feb 2010
Posts: 5
Newbie
|
Newbie
Joined: Feb 2010
Posts: 5 |
anyone know what the part number for the piston arm is? or where i can get some? The bearing on the end has worn causing noisy running. many thanks
|
|
|
|