Hi folks

I'm after some information regarding the testing of nebulisers. As i haven't got access to any devices or manuals i'm after those of you folks who have the knowledge and what sort of pressure and flow testing you carry out.

Also what test equipment do you use for the servicing of these items and are there any specifics that might be useful. Im looking at the Medix range like the AC4000.

Any information what be of great help folks.

darran1965 #23977 13/06/07 11:26 AM
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72
Yes Darran, output pressure and flow are the final tests to be carried out once you have cleaned out all the crud, changed the filter, confirmed electrical safety etc. There used to be a well-known test rig (consisting of a pressure gauge, and flowmeter, mounted on a board) that I've come across in many biomed shops. I think they were supplied by Medic-Aid (in those days), or maybe it was Medix (someone out there will know)? Once you have sorted out your test rig (and you could always make your own), you will need a little chart listing the go/no-go limits for any and all nebs you service. Again, perhaps one of the Brethren may be kind enough to offer such a list right here! smile

Last edited by Geoff Hannis; 13/06/07 11:35 AM. Reason: It needed editing.

If you don't inspect ... don't expect.
Geoff Hannis #23980 13/06/07 12:17 PM
Joined: Sep 2005
Posts: 70
Likes: 9
kit Offline
Scholar
Offline
Scholar
Joined: Sep 2005
Posts: 70
Likes: 9
Considering you can pick up a nebuliser for £60 I would be inclined to do minimal work on these units. All nebulisers I have are checked at a pressure of 20PSI. If the flow corresponds to what the manufactures state I would check the plug, fuse, mains cable and ensure it is not physically damged. If the unit is more than 3 years old I would scrap it immediately if it has a low output, is physically damaged or even if it is particulary dirty. I wouldn’t consider going inside a particularly dirty unit and cleaning it, as I wouldn’t consider that economical.

Last edited by kit; 13/06/07 1:18 PM. Reason: Grammar
kit #23981 13/06/07 12:43 PM
Joined: Jan 2006
Posts: 141
Expert
Offline
Expert
Joined: Jan 2006
Posts: 141
AC4000 nebulisers specified dynamic flow rate is at least 7 lpm @ 15 psi.

Medic Aid ( now Respironics ) supplies the testing rig


Joined: Jul 2000
Posts: 1,969
Likes: 32
Hero
Offline
Hero
Joined: Jul 2000
Posts: 1,969
Likes: 32
Takes 1 minute to check on a test rig. Some repairs are still worth doing, but I agree it should be limited to no more than 10 - 15 mins or throw it in the bin.We have an assitant technician that does nebs (changes filters, cracked cases, etc).


Be Proactive and reactive.
Joined: May 2007
Posts: 8
Newbie
OP Offline
Newbie
Joined: May 2007
Posts: 8
Thanks for the response guys i wasn't aware of the that you could actually buy a test rig. I've not been able to find this on the web with the info you guys gave so if anyone could supply a link that would be great. Also as anyone got a list they use of the limits for nebs when testing. Anyone got a contact for things like filters and such like for Medix? as i've had no luck smile


darran1965 #23997 14/06/07 7:22 AM
Joined: Aug 2001
Posts: 797
Likes: 1
KM Offline
Philosopher
Offline
Philosopher
Joined: Aug 2001
Posts: 797
Likes: 1
DARRAN
MEDIX IS CLEMENT CLARKE, ASK FOR THEIR REPIRATORY PRODUCTS TEAM FOR HOSPITALS (ITS A DIFFERENT TEAM TO JOE PUBLIC).
MEDIC AID IS RESPIRONICS ASK FOR TECHNICAL SUPPORT DEPARTMENT.
IN BOTH CASES THEY ARE USUALLY VERY APPROACHABLE AND CAN USUALLY SORT ANY PROBLEMS.
WE HAVE 2 TEST RIGS THAT WE BOUGHT IN 1995 FROM THE ABOVE AT £135.00 EACH. THEY ARE BASICALLY A FLOMETER AND GAUGE MOUNTED ON THE PIECE OF BOARD. YOU CONNECT THE NEB OUTLET TO A PIECE OF TUBE AND GET A READING ON BOTH.


KM #23998 14/06/07 7:36 AM
Joined: Jan 2006
Posts: 141
Expert
Offline
Expert
Joined: Jan 2006
Posts: 141
KM - Don't shout here, we ain't deaf on this forum! grin

Darran - check your PMs

Last edited by Jonathan Wells; 14/06/07 8:14 AM.
Joined: May 2002
Posts: 137
Expert
Offline
Expert
Joined: May 2002
Posts: 137
Whilst the pressure guage is a standard item, the flow meter needs a calibrated nozzle in series with it as the flow (current) and pressure (voltage) specifications are stated with the correct flow restriction (resistance).

RICK #48025 06/08/10 12:27 PM
Joined: Feb 2010
Posts: 5
Newbie
Offline
Newbie
Joined: Feb 2010
Posts: 5
anyone know what the part number for the piston arm is? or where i can get some? The bearing on the end has worn causing noisy running. many thanks


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
2 members (daisizhou, RDS@MES), 21,083 guests, and 80 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,288
Posts74,594
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5