#3474 07/10/05 2:52 PM
Joined: May 2004
Posts: 22
Dreamer
OP Offline
Dreamer
Joined: May 2004
Posts: 22
I have tried seaching for the answer to this question, but have not had much luck.

Why should we use a manufacturers cuffs.

In other words is there a difference between the cuffs that means that the readings would be inacurate if we used a "rouge" cuff? Is it the material, or its flexibilty that affects the NIBP reading, or alters the data being given to the machines "algorithms"

Any help gratefully received.


The "answer" is of course 42
#3475 07/10/05 3:21 PM
Joined: Mar 2004
Posts: 47
Technologist
Offline
Technologist
Joined: Mar 2004
Posts: 47
As long as a two hose cuff is used on a two hose machine and the connectors are the correct ones I don't see any problems in using third party cuffs.

#3476 09/10/05 5:20 AM
Joined: Sep 2005
Posts: 71
Scholar
Offline
Scholar
Joined: Sep 2005
Posts: 71
on many schiller argus monitors, ge solar 8000 and innomed monitors, even HP and philips we used third party cuffs with no problem. sometimes they even lasted longer than those supplyed by the manufacturer. as long the size and the principle of measurement is the same (mounted mic. or oscilometry).
you can give it a test anyway
cheers

#3477 10/10/05 9:31 AM
Joined: May 2004
Posts: 7
Newbie
Offline
Newbie
Joined: May 2004
Posts: 7
The problem is that most manufacturers will state that their product has been validated with a particular cuff and consequently if used with a generic cuff the validation is no longer valid.

This is I am sure is in the great scheme of things a load of crap, but if you are standing up in a court room and have to swear that a bit of nibp kit was routinely checked as per the maufacturers spec but you were using a spurious cuff then a smart arse lawyer will hang you out to dry...

Life gets tedious don't it.....


"Never eat more than you can lift"

Miss Piggy 1982ish
#3478 10/10/05 10:05 AM
Joined: Jul 2002
Posts: 2,020
Hero
Offline
Hero
Joined: Jul 2002
Posts: 2,020
I had a large monitor manufacturer trying to sell me their range of cuffs for use with other monitors. He pointed out that the other monitor manufacturers are unlikely to actually produce their own consumables but his company did produce cuffs and had probably produced the ones that the other manufacturers sold so it was perfectly alright to use his cuffs on any monitor.
If you look at a lot of the cuffs on the market, especially the third party ones, they do look remarkably similar.
Robert


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
#3479 10/10/05 10:58 AM
Joined: Jun 2003
Posts: 12
Novice
Offline
Novice
Joined: Jun 2003
Posts: 12
We have been using third party cuffs from Henley's for a couple of years without a problem.
Before deciding to change over we tested the cuffs and found no difference in accuracy compared to original manufacturers cuffs.
With all the different cuff sizes to stock it's alot easier than storing several makes and they only cost £9 each.
Some manufacturers would advise against using other cuffs but if they give the same accuracy when tested, why not use them? Could they just be protecting their sales of consumables?

#3480 10/10/05 11:49 AM
Joined: May 2004
Posts: 39
Visionary
Offline
Visionary
Joined: May 2004
Posts: 39
Interesting, manufacturers have always insisted using their own cuffs on their equipment, however since GE took over Critikon they now supply Criticon cuffs for what was marquette equipment where privously this would have been frowned on.
I wonder whether they supply the same cuffs for Datex monitors? confused

#3481 10/10/05 11:58 AM
Joined: Dec 2003
Posts: 282
Master
Offline
Master
Joined: Dec 2003
Posts: 282
Funny you should raise the question Lee.
The current Datex Adult cuffs have CRITIKON on them with a slightly different part number, probably due to the different connectors!
Graham

#3482 10/10/05 2:13 PM
Joined: Dec 2003
Posts: 105
BK Offline
Savant
Offline
Savant
Joined: Dec 2003
Posts: 105
Just to put another angle on this. The NPSA is recommending Trusts move towards latex-free environment because of the increase in latex allergy cases. So you may want to consider this when deciding on supplier.
Soon there will be a requirement for manufacturers to state (on the packaging)if products contain latex.

#3483 11/10/05 1:17 PM
Joined: Jul 2002
Posts: 2,020
Hero
Offline
Hero
Joined: Jul 2002
Posts: 2,020
I was looking through the Artemis catalogue and noticed they sell CAS medical cuffs (which are latex free). It says in a bold box "CAS Medical's cuffs are certified for use with all major manufacturer's monitors".
Who certifies this? Or are CAS Medical just taking responsibility? Does this mean we can use this third party cuffs and refer any accuracy problems the lawyers might ask of us back to the maufacturer?

Several major monitor manufacturers advertise the fact that they use a certain branded NIBP technology. Does this mean they use the same cuffs with their own brand on? I presume so if they are purchasing the whole technology. So much for "must use only their cuffs".

Robert


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 5,426 guests, and 79 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,288
Posts74,594
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5