Joined: Oct 2003
Posts: 51
Scholar
Offline
Scholar
Joined: Oct 2003
Posts: 51
We here run the Aquarius and have not had this problem. But we have had exactly the same symptoms on our heart bypass pumps.

The general source of this interferance seems to be as you say above micro currents being generated on the tube surface or within the blood inside the tubing, caused by electro-static build up.

2 quick things you could check. 1) the roller occulsion of the pumps on to the tubing itself - if it is set slightly high so as to over occulude the pump then this can stress the tubing leading to an electo-static problem. 2) I know this sounds daft but try spraying the pump heads and raceways with a little anti-static spray.

Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

Yes, anti-static spray could well be the answer here. So let's hope we hear good news soon! smile


If you don't inspect ... don't expect.
Joined: Jan 2005
Posts: 768
Philosopher
Offline
Philosopher
Joined: Jan 2005
Posts: 768
As Paul points out, it is a well known phenomenon with Heart-Lung Bypass machines due to static build up on the plastic pump raceway. Anti-static spray can help alleviate this.
However, the pumps on CVVH machines tend to have pre-set occlusion and are non-adjustable and also it is impracticable to keep spraying with anti-static.
Make sure that the patient's ECG electrodes are correctly applied (as pointed out by other contributors) as this will alleviate the problem in most cases. Also check that the CVVH machine pumps have been lubricated correctly with regard to the rollers and main shaft.


Sometimes You Can't Make It On Your Own.
Joined: Nov 2008
Posts: 1
Newbie
Offline
Newbie
Joined: Nov 2008
Posts: 1
we have this problem and find that a quick spray of antistatic into the blood pump pathway reduces the interference.

Joined: Oct 2007
Posts: 4
STEVEA Offline OP
Newbie
OP Offline
Newbie
Joined: Oct 2007
Posts: 4
Thanks guys for all your suggestions, I'll try some of these out next time we experience the problem. I've already tried connecting the equipotential points on the machines but this made no difference.

Joined: Sep 2004
Posts: 213
Master
Offline
Master
Joined: Sep 2004
Posts: 213
How about using diathermy proof ECG leads, they cut down on the interference cause by ESU.

It maybe worth a try as you may already have these leads in theatre.

A.M smile


Barry

Be not afraid of greatness; some are born great, some achieve greatness, and others have greatness thrust upon them
Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71

Whenever possible, it's always best to treat the cause, rather than the symptoms! smile


If you don't inspect ... don't expect.
Joined: Sep 2004
Posts: 213
Master
Offline
Master
Joined: Sep 2004
Posts: 213
Just thought it might be easier than wrapping patient treatment hose in tin foil.

How about sticking an anti static strap on the patient.

A.M smile


Barry

Be not afraid of greatness; some are born great, some achieve greatness, and others have greatness thrust upon them
Joined: Feb 2004
Posts: 14,798
Likes: 71
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,798
Likes: 71
Originally Posted By: Aston Martin
Just thought it might be easier than wrapping patient treatment hose in tin foil.

Perhaps ... but where's the fun in that? wink


If you don't inspect ... don't expect.
Page 2 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 5,907 guests, and 21 robots.
Key: Admin, Global Mod, Mod
Newest Members
Chris 11, j9_PLC, nece, Vitya, Shenzhen007
10,358 Registered Users
Forum Statistics
Forums26
Topics11,248
Posts74,481
Members10,358
Most Online49,431
11 hours ago
Powered by UBB.threads™ PHP Forum Software 7.7.5