Dear All.
One of our CAS MEDICAL OSCILLOMATE 9300 NIBP monitor is restarting with MESSAGES " RETRYING" after 3 to 5 seconds "FAILED" message appears. But still restarting.
Any one have a hunch or idea what may be causing this.
Waititng for positive reply.
Regards

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Sounds like there's a leak there somewhere, Mate! Hopefully, it will just be something simple, like the cuff or its tubing. smile


If you don't inspect ... don't expect.
Joined: Nov 2008
Posts: 12
Novice
Offline
Novice
Joined: Nov 2008
Posts: 12
Replaced the cuff and hose and confirm there is no leak, still error is persisting check internal leak
thanks
rahim

Joined: Apr 2005
Posts: 380
Laiq Offline OP
Sage
OP Offline
Sage
Joined: Apr 2005
Posts: 380
thanks geoff, & abdul.
U are right, the pump is not working at all during bootup.
Checked the pump driver ckt it is ok, FET & AND GATE driver.
The processor is not giving PUMP ON command.
I think there is some internal calibration error thats why it is not proceeding to normal operation mode.
Also i checked the supplies they look ok to me.
If i can get a starting point i can isolate the problem insha ALLAH.
Jaza-k-ALLAH for your help freinds.

Joined: Dec 2007
Posts: 18
Novice
Offline
Novice
Joined: Dec 2007
Posts: 18
Hi Laiq, this seems to indicate a transducer problem to me. I think I would first check that the unit reads pressure OK, now that you have established that it is not an air leak. If the pressure does not rise then you have found your problem. All automated nibp monitors have some logic that monitors cuff pressure and if it does not rise sufficiently during a reading it will come up with an error. You may want to speak to CAS directly and their contact details are on the web www.casmed.com.

Richard


Declaration of Interest: SunTech Medical is a NiBP device manufacturer. Participation in this forum is solely to seek opinion and feedback from, and offer assistance to the EBME community.
Joined: Apr 2005
Posts: 380
Laiq Offline OP
Sage
OP Offline
Sage
Joined: Apr 2005
Posts: 380
Dear Richie
Thanks for your reply. I have managed to repair one unit successfully. Yes it is transducer & its associated ckt fault.
Processor is not getting the required transducer readings in order to proceed to normal operation.
Now currently working on last unit to make it work again.
CAS medical have replied to me as one of my freind did contacted them on my behalf. The response was same as i was expecting, that this model is obsolete, no diagram or data for this model, replace it with new one :-)
Thanks anyway for viewing this post & replying.
Regards
LAIQ


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (Huw), 8,928 guests, and 319 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,287
Posts74,592
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5