Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rating: 5
Page 3 of 7 1 2 3 4 5 6 7
Joined: Feb 2002
Posts: 834
Philosopher
Offline
Philosopher
Joined: Feb 2002
Posts: 834
Agreed! EST is necessary! This is a golden rule! But again is it necessary to record ALL the results or have a simple EST pass/fail on the service sheet? We treat our lab equipment as PAT only (mainly because it is!) For PAT we just record Pass/Fail! All our accreditors including CAPS and JCI accept this record! Would you say this is acceptable for Medical Equipment?

Eddie

Joined: Sep 2003
Posts: 260
Master
Offline
Master
Joined: Sep 2003
Posts: 260
Its worth bearing in mind that i myself and i suspect many of you have witnessed equipment in lethal condition happily pass EST with flying colours, one item had actually gave a nurse an electric shock but still sailed through EST. What these occurrences boil down to is good old physical inspection, can't beat it. With this in mind, i don't see the point in comparing previous EST's. I do however agree that EST is still necessary as it does test the "unseen" i.e. the internal electrical safety of the device. I think a simple pass / fail is all that need be recorded on the service report. Personally i staple the printout to the sheet where as others discard it which seems a waste.
Merry Xmas
Topper

Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

I can feel a poll (or two) coming on! Huw? smile

What EST results should be saved:-

1) None
2) Pass/Fail only
3) Full results
4) Depends upon type of device
5) As required by contract
6) As required for accreditation

EST results should be saved (1):-

1) In the tester only
2) On printouts (fade away)!
3) On printouts clipped to device records
4) Written on job sheet/PM report only
5) On computer in raw format
6) On computer in processed format
7) Combination of written and computer
8) As required by contract
9) As required for accreditation

EST results should be saved (2):-

1) Written in a log-book
2) Written in device records
3) Filed with job sheet/PM report
4) Entered into computer database
5) Downloaded into computer
6) As required by contract
7) As required for accreditation

The previous EST result(s) should be referred to:-

1) Never
2) Always
3) When there is doubt
4) Depends upon type of device
5) As required by contract

EST results should be analyzed (1):-

1) Never (waste of time)
2) When similar devices fail EST
3) As required by management
4) Depends upon type of device
5) As required by contract

EST results should be analyzed (2):-

1) Never (see above)
2) Manually following search criteria
3) By computer following search criteria
4) As required by contract
5) During accreditation if asked for

... something like that? smile


If you don't inspect ... don't expect.
Joined: Jun 2000
Posts: 2,412
Likes: 13
Huw Online Content
Hero
Online Content
Hero
Joined: Jun 2000
Posts: 2,412
Likes: 13
Go ahead Geoff, I've modified the permissions to allow polls in topic responses (as opposed to only at creation of the topic).

Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

OK, cheers for that.

Perhaps we had better wait first for poll suggestions from the Three (or so) Wise Men! smile


If you don't inspect ... don't expect.
Joined: Feb 2002
Posts: 834
Philosopher
Offline
Philosopher
Joined: Feb 2002
Posts: 834
Lets go for it! I am interested to see what the general consensus is!

Just go for Poll "A" first mate!

What EST results should be saved:-

1) None
2) Pass/Fail only
3) Full results
4) Depends upon type of device
5) As required by contract
6) As required for accreditation

Ed

Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

Patience, young Eddie, patience! At least wait until I've been across the road for a bacon sandwich, and a cup of tea. Come to think of it, I might even pick up a turkey sandwich ... I have to make an effort to enter into the spirit of things, you know! wink


If you don't inspect ... don't expect.
Joined: Feb 2002
Posts: 834
Philosopher
Offline
Philosopher
Joined: Feb 2002
Posts: 834
I find a good old Malt does the trick rather nicely old bean!

Eddie

Joined: Feb 2004
Posts: 14,814
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,814
Likes: 72

... sherry was being offered to customers at the sandwich shop. All very civilized. But, there again, that's South Yorkshire for you. Traditional values! wink


If you don't inspect ... don't expect.
Joined: May 2001
Posts: 457
Sage
Offline
Sage
Joined: May 2001
Posts: 457
Please give the full title Tha Knows !! its " The Peoples Democratic Socialist Republic of South Yorkshire " appen sithee.

Page 3 of 7 1 2 3 4 5 6 7

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (Huw), 7,710 guests, and 51 robots.
Key: Admin, Global Mod, Mod
Newest Members
Czr, Jimmy Li, AndrewS1, Lam, Danny1621
10,388 Registered Users
Forum Statistics
Forums26
Topics11,274
Posts74,579
Members10,388
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5