|
|
|
|
Joined: Jul 2007
Posts: 29
Dreamer
|
OP
Dreamer
Joined: Jul 2007
Posts: 29 |
hello dear all! i have a problem wit a PB companion 492A oxygen concentrator: it only deliver 40% o² at 4L/mn.then afer 2 or 3 minutes it shut down with service alarm on . the compressor seems to be ok but i wonder how to test the valves or if it is "safe" to open them , according to the age of the machine (1996). does some one has some workshop tips about it ? i have also problems with old Devilbiss 505 , no O² flow at all , compressor good , no idea of what to check first.
thanks for reading
jean-michel
|
|
|
|
|
Joined: Oct 2008
Posts: 2
Newbie
|
Newbie
Joined: Oct 2008
Posts: 2 |
Hi mate spent years fixing devilbiss LT and 515 series concentrators,The change over valves and fans on these models are the main thing to go on these . It is easy to remove the change over valve and clean the pistons in side.Hope this helps if you want any more help drop me a line
Last edited by kab; 09/01/09 3:45 PM.
|
|
|
|
|
Joined: Jul 2007
Posts: 29
Dreamer
|
OP
Dreamer
Joined: Jul 2007
Posts: 29 |
thank you Kab for your advices! i have open one valve to seen how it look inside and i found that the rubber membranes had become hard. i don't know if a service kit is still availlable for those rather olds machines , but i will try to find some.
best regards
jean-michel
|
|
|
|
|
Joined: Feb 2004
Posts: 14,802 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,802 Likes: 72 |
Is this .pdf manual any use? 
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Jul 2007
Posts: 29
Dreamer
|
OP
Dreamer
Joined: Jul 2007
Posts: 29 |
thanks my dear , but unfortunately this is not the good one. but now the oxygen plant is working well , so we don't need smalls concentrators anymore. thank you Geoff for your searchs !
best regards
jean michel
|
|
|
|
1 members (Diego Lins),
6,265
guests, and
34
robots. |
|
Key:
Admin,
Global Mod,
Mod
|
|
|
Forums26
Topics11,255
Posts74,507
Members10,364
| |
Most Online59,530 Apr 30th, 2026
|
|
|
|
|
|