Hi All. I'm trying to get some feedback on how hospitals handle infection control of NiBP cuffs. Can anyone give me a pointer to some guidelines?
Thanks
Richard
Hi Richard,
From our Infection Control Cleaning of Reusable Equipment document - "Any reusable equipment or equipment that comes into contact with intact skin must be cleaned before it is used. Hot water and detergent is acceptable for routine cleaning.
If a surface is contaminated by blood or other body fluids disinfection is required."
To quote from our Infection Control Standard Precaution document - "Patient care equipment which becomes contaminated with infective material, must be either discarded properly, or correctly disinfected before re-use."
---------------
The following is a bit off topic ...
we normally replace our BP cuffs not because of infection control but because the hook side of the velcro doesnt attach well to the loop side. It appears to be still physically intact but we know its time to change it when initial pressure goes to 180 or above before the machine starts to bleed air. You can imagine how the patient reacts.
The bladders are ok, its just the velcro, and our cuffs bear the bp machine manufacturer's logo. We confirmed that it was the velcro as we "auditioned" several cuffs from Advantage Med and these lasted far longer than the "original".