hi. i need help. how to unlock 'data lock' on this perfusor.

Joined: Feb 2005
Posts: 2
Newbie
Offline
Newbie
Joined: Feb 2005
Posts: 2
Hi Jak,

to remove the data lock hold go into the special functions menu (SF), and go to the data lock option. Hold down the decimal point key, then press the 'off' soft key.

Hope this helps

Joined: May 2009
Posts: 32
jak Offline OP
Visionary
OP Offline
Visionary
Joined: May 2009
Posts: 32
thank andy. actually my unit had this 'with data lock' message and i can't do anything even go to data lock option in SF menu. any idea?

Joined: Feb 2005
Posts: 2
Newbie
Offline
Newbie
Joined: Feb 2005
Posts: 2
Hi Jak,

you can only get to the SF menu when the pump is not running..

Does the SF key work (keep pressing it with the pump stopped to go through the options)?

Does the pump work correctly in other ways, syringe loading, start / stop (data lock only locks the rate change and bolus)?

If data lock is not in the SF menu then if you put in the service key it turns all the SF functions on (just while the key is in), so that you can turn data lock off.

Is it the PCA version? (Says PCA on the front) To unlock these you need a PCA key (dongle for the back)...

Good luck.

Joined: May 2009
Posts: 32
jak Offline OP
Visionary
OP Offline
Visionary
Joined: May 2009
Posts: 32
hi andy. lets say,

First, i put in the 10 ml B-D PLpak syringe, then press ON to prime it 1 ml. There are no option key to select syringe type. Only ON and END. then i press END soft key.

Then the display shows SF softkey only and 'with data lock' message. When i keep pressing SF softkey, it goes to dosage calculation , bolus mode pressure limit and battery capacity and no option softkey to change any parameters each. lastly it goes back to 'with data lock' message.
The flow rate also can't change. it still shown 000.0 ml/h. thus, the 'without rate' alarm appeared.

likely, i need to go to service mode to 'off' the data lock right?


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
3 members (Spidey, jacks, daisizhou), 3,781 guests, and 99 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,288
Posts74,594
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5