We experience a recurrent problem with our PCA pumps - during operation the pump would beep for about 1 second without any warning message. Normal operation isnt interrupted, and this would sometimes recur within the same session. It happens to many devices, many times. Alaris service gave a vague opinion that this is normal behavior between demands but other pumps of this model show no such behavior. Has anyone encountered this ? found a cure ? Thanks David
|
|
|
|
|
Joined: Dec 2003
Posts: 282
Master
|
Master
Joined: Dec 2003
Posts: 282 |
David The bad news is we have 15 of these PCAs and have never come across this problem Graham
|
|
|
|
|
Joined: Jul 2004
Posts: 89
Adept
|
Adept
Joined: Jul 2004
Posts: 89 |
Hello David,
Are you not confusing the "pump working on battery" bleep, it sounds very similiar to your problem?
Tracy
Self Employed Governance & Medical Devices Consultant / NHS Professional - Medical Engineer & Medical Devices Trainer
|
|
|
|
|
Joined: Oct 2005
Posts: 8
Newbie
|
OP
Newbie
Joined: Oct 2005
Posts: 8 |
Tracy The devices are constantly plugged in to AC. But you did give me an idea - could it "autonomously" switch between operation on AC and Battery ? I've tried to simulate this by quickly disconnecting/reconnecting the AC plug on the pump side but I then get the "AC Fail" alarm loud and clear and not a fleeting beep. Problem is the departments keep sending the devices to me as "broken - beeps" and all I can do is verify the claims, send it to Alaris and get it back same as it went. Thanks David
|
|
|
|
|
Joined: Dec 2004
Posts: 135
Expert
|
Expert
Joined: Dec 2004
Posts: 135 |
David, I don't know if this will help? I seem to remember that these pumps will beep like this when run on battery and if the 'chirp low alarms' is configured on in the general options. To try and determine if this is a factor, hold the start button whilst switching on then enter the code 251 and go to option 2 'general options'. 'Chirp Low Alarms' is number 7 in the general options, if yes is selected as an option, as Tracy mentioned, this could be what you're experiencing
Regards,
Ivor
|
|
|
|
|
Joined: Sep 2004
Posts: 213
Master
|
Master
Joined: Sep 2004
Posts: 213 |
I have this same problem with one of our P5000 PCAM pump. We were recommend to change the ribbon cable between PSU board and the main Processor board. If did not work try fitting a new PSU board. This cured my problem. Be aware there has been a MHRA Medical Device Alert MDA/2005/068 issued on the 08.12.2005. This is to do with the battery run time and alerts on all “P” series pumps. A.M 
Barry
Be not afraid of greatness; some are born great, some achieve greatness, and others have greatness thrust upon them
|
|
|
|
|
Joined: Oct 2003
Posts: 89
Adept
|
Adept
Joined: Oct 2003
Posts: 89 |
Originally posted by Shins: David, I don't know if this will help? I seem to remember that these pumps will beep like this when run on battery and if the 'chirp low alarms' is configured on in the general options. To try and determine if this is a factor, hold the start button whilst switching on then enter the code 251 and go to option 2 'general options'. 'Chirp Low Alarms' is number 7 in the general options, if yes is selected as an option, as Tracy mentioned, this could be what you're experiencing
Regards,
Ivor Great idea. Why don't we just put all the service codes on the website for nurses to learn. Would a private message not have been a better idea?
It's not something you can teach. I am that damn good.
|
|
|
|
|
Joined: Dec 2004
Posts: 135
Expert
|
Expert
Joined: Dec 2004
Posts: 135 |
Unfortunately all these codes are available in the Alaris User guides............. 
|
|
|
|
|
Joined: Oct 2003
Posts: 89
Adept
|
Adept
Joined: Oct 2003
Posts: 89 |
Must be working from a different one from me then because the latest P5000 DFU just states "for technician access codes please refer to the technical service manual" It shows them the procedure but doesn't give the code
It's not something you can teach. I am that damn good.
|
|
|
|
|
Joined: Dec 2004
Posts: 135
Expert
|
Expert
Joined: Dec 2004
Posts: 135 |
From what I've been told I think that they've been made aware of the amount of information in the P series user guides being perhaps a little too useful. You're right though, maybe we spend so much time having to watch out for big brother that we sometimes forget about little sister....... 
|
|
|
|
1 members (daisizhou),
3,658
guests, and
121
robots. |
|
Key:
Admin,
Global Mod,
Mod
|
|
|
Forums26
Topics11,289
Posts74,595
Members10,395
| |
Most Online69,877 Jul 17th, 2026
|
|
|
|
|
|