We experience a recurrent problem with our PCA
pumps - during operation the pump would beep
for about 1 second without any warning message.
Normal operation isnt interrupted, and this
would sometimes recur within the same session.
It happens to many devices, many times.
Alaris service gave a vague opinion that this
is normal behavior between demands but other
pumps of this model show no such behavior.
Has anyone encountered this ? found a cure ?
Thanks
David

Joined: Dec 2003
Posts: 282
Master
Offline
Master
Joined: Dec 2003
Posts: 282
David
The bad news is we have 15 of these PCAs and have never come across this problem
Graham

Joined: Jul 2004
Posts: 89
Adept
Offline
Adept
Joined: Jul 2004
Posts: 89
Hello David,

Are you not confusing the "pump working on battery" bleep, it sounds very similiar to your problem?


Tracy


Self Employed Governance & Medical Devices Consultant /
NHS Professional - Medical Engineer & Medical Devices Trainer
Joined: Oct 2005
Posts: 8
DavidJ Offline OP
Newbie
OP Offline
Newbie
Joined: Oct 2005
Posts: 8
Tracy
The devices are constantly plugged in to AC.
But you did give me an idea - could it
"autonomously" switch between operation on
AC and Battery ? I've tried to simulate this
by quickly disconnecting/reconnecting the AC
plug on the pump side but I then get the
"AC Fail" alarm loud and clear and not a
fleeting beep.
Problem is the departments keep sending the
devices to me as "broken - beeps" and all I
can do is verify the claims, send it to Alaris
and get it back same as it went.
Thanks
David

Joined: Dec 2004
Posts: 135
Expert
Offline
Expert
Joined: Dec 2004
Posts: 135
David, I don't know if this will help?
I seem to remember that these pumps will beep like this when run on battery and if the 'chirp low alarms' is configured on in the general options.
To try and determine if this is a factor, hold the start button whilst switching on then enter the code 251 and go to option 2 'general options'.
'Chirp Low Alarms' is number 7 in the general options, if yes is selected as an option, as Tracy mentioned, this could be what you're experiencing

Regards,

Ivor

Joined: Sep 2004
Posts: 213
Master
Offline
Master
Joined: Sep 2004
Posts: 213
I have this same problem with one of our P5000 PCAM pump. We were recommend to change the ribbon cable between PSU board and the main Processor board. If did not work try fitting a new PSU board. This cured my problem.

Be aware there has been a MHRA Medical Device Alert MDA/2005/068 issued on the 08.12.2005. This is to do with the battery run time and alerts on all “P” series pumps.

A.M smile


Barry

Be not afraid of greatness; some are born great, some achieve greatness, and others have greatness thrust upon them
Joined: Oct 2003
Posts: 89
Adept
Offline
Adept
Joined: Oct 2003
Posts: 89
Quote:
Originally posted by Shins:
David, I don't know if this will help?
I seem to remember that these pumps will beep like this when run on battery and if the 'chirp low alarms' is configured on in the general options.
To try and determine if this is a factor, hold the start button whilst switching on then enter the code 251 and go to option 2 'general options'.
'Chirp Low Alarms' is number 7 in the general options, if yes is selected as an option, as Tracy mentioned, this could be what you're experiencing

Regards,

Ivor
Great idea. Why don't we just put all the service codes on the website for nurses to learn.

Would a private message not have been a better idea?


It's not something you can teach. I am that damn good.
Joined: Dec 2004
Posts: 135
Expert
Offline
Expert
Joined: Dec 2004
Posts: 135
Unfortunately all these codes are available in the Alaris User guides............. smilewink

Joined: Oct 2003
Posts: 89
Adept
Offline
Adept
Joined: Oct 2003
Posts: 89
Must be working from a different one from me then because the latest P5000 DFU just states "for technician access codes please refer to the technical service manual"
It shows them the procedure but doesn't give the code


It's not something you can teach. I am that damn good.
Joined: Dec 2004
Posts: 135
Expert
Offline
Expert
Joined: Dec 2004
Posts: 135
From what I've been told I think that they've been made aware of the amount of information in the P series user guides being perhaps a little too useful.
You're right though, maybe we spend so much time having to watch out for big brother that we sometimes forget about little sister.......
eek

Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 3,658 guests, and 121 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,289
Posts74,595
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5