Originally posted by Tracy:
Hello David
As previously mentioned by Aston, the interconnect cable on these pumps can also be a problem.
The interconnect cable , in my experience does not last that long. (the one above the transformer is the one i refer to)
Alaris or whatever they are called where you are, supply replacecements, and the new ones are more better than the old ones. They fit better and the creases are in the right place as to avoid the heat from the transformer as much as possible. I think they cost a few pounds each, but in my experience of looking after these pumps, the new connector was superior.
A very common fault with the connector is an error-onous low battery alarm.
I found that with the case open problems with the interconnect cable could be easily reproduced by moving the cable.
The engineers at alaris in uk are a very helpful bunch so ask them!!!
Tracy
Hi Tracy
Finished checking the chirp alarm on all pumps - all were "no". Apparently this was the untamperedwith default.
Tried playing with the cable as you suggested,
both on a good and a bad device - nothing !
no beeps, no alarms -nothing.
Switched the PS between the good and the bad
units - lo and behold the problem appears now
in BOTH pumps. Switching the PS back, now does
not solve the problem in the previously good
pump.
Will post further.
Thank you people for the cooperation.
David