Home Articles Downloads Forum Products Services EBME Expo Contact ☰
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 2 of 2 1 2
Joined: Oct 2005
Posts: 8
DavidJ Offline OP
Newbie
OP Offline
Newbie
Joined: Oct 2005
Posts: 8
Quote:
Originally posted by Shins:
David, I don't know if this will help?
I seem to remember that these pumps will beep like this when run on battery and if the 'chirp low alarms' is configured on in the general options.
To try and determine if this is a factor, hold the start button whilst switching on then enter the code 251 and go to option 2 'general options'.
'Chirp Low Alarms' is number 7 in the general options, if yes is selected as an option, as Tracy mentioned, this could be what you're experiencing

Regards,

Ivor
Definitely possible. I'm gonna check this on all
our PCAs - we have some 50 of them all over the
place...
Thanks Ivor !

Joined: Jul 2004
Posts: 89
Adept
Offline
Adept
Joined: Jul 2004
Posts: 89
Hello David

As previously mentioned by Aston, the interconnect cable on these pumps can also be a problem.

The interconnect cable , in my experience does not last that long. (the one above the transformer is the one i refer to)

Alaris or whatever they are called where you are, supply replacecements, and the new ones are more better than the old ones. They fit better and the creases are in the right place as to avoid the heat from the transformer as much as possible. I think they cost a few pounds each, but in my experience of looking after these pumps, the new connector was superior.

A very common fault with the connector is an error-onous low battery alarm.

I found that with the case open problems with the interconnect cable could be easily reproduced by moving the cable.

The engineers at alaris in uk are a very helpful bunch so ask them!!!

Tracy


Self Employed Governance & Medical Devices Consultant /
NHS Professional - Medical Engineer & Medical Devices Trainer
Joined: Oct 2005
Posts: 8
DavidJ Offline OP
Newbie
OP Offline
Newbie
Joined: Oct 2005
Posts: 8
Quote:
Originally posted by Tracy:
Hello David

As previously mentioned by Aston, the interconnect cable on these pumps can also be a problem.

The interconnect cable , in my experience does not last that long. (the one above the transformer is the one i refer to)

Alaris or whatever they are called where you are, supply replacecements, and the new ones are more better than the old ones. They fit better and the creases are in the right place as to avoid the heat from the transformer as much as possible. I think they cost a few pounds each, but in my experience of looking after these pumps, the new connector was superior.

A very common fault with the connector is an error-onous low battery alarm.

I found that with the case open problems with the interconnect cable could be easily reproduced by moving the cable.

The engineers at alaris in uk are a very helpful bunch so ask them!!!

Tracy
Hi Tracy
Finished checking the chirp alarm on all pumps - all were "no". Apparently this was the untamperedwith default.
Tried playing with the cable as you suggested,
both on a good and a bad device - nothing !
no beeps, no alarms -nothing.
Switched the PS between the good and the bad
units - lo and behold the problem appears now
in BOTH pumps. Switching the PS back, now does
not solve the problem in the previously good
pump.
Will post further.
Thank you people for the cooperation.
David

Joined: Dec 2004
Posts: 135
Expert
Offline
Expert
Joined: Dec 2004
Posts: 135
David, the cable which we've had trouble with is the one that Aston Martin refers to.
This is the 16 way ribbon cable which interconnects Power Supply PCB and Control PCB, I believe that the part number is 1000-EL-00135.

Ivor

Page 2 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 3,658 guests, and 121 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,289
Posts74,595
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5