We use a Veri-Cal Pressure Transducer Tester from
Utah Medical which might fit the bill.
It is basically a pressure/vaccumm gauge which has a large knob on the side winding this in or out adjusts the pressure.
Spec follows:-
Veri-Cal™ (650-900)
Operating Pressure Range ±300 mmHg, 1 mmHg resolution
±199.9 mmHg, 0.1 mmHg resolution
Operating Temperature 15° C to 40° C (60° F to 105° F)
Storage Temperature -20° C to 65° C (-5° F to 150° F)
Operating Humidity 80% RH Max., Non-Condensing
Accuracy ±1% of reading or ±1 mmHg, whichever is greater
Repeatability and Hysteresis ±0.15% full scale output
Overpressure 750 mmHg.