We were recently doanted some Pari Junior Boy nebulisers (new and boxed), and upon inspection found that the live and neutral spade connectors inside the plastic case were not insulated. The device is CE marked to MDD and marked as double insulated. It seems to me that the only protective barrier against electric shock is the plastic case itself.
I made equiries of the MDA and was told the following:
"An air gap around conductive parts counts as insulation"!!
I can't see how an air gap could prevent electric shock in the single fault condition of a broken or damaged case.
Has anyone else experienced this with a medical device? All other classII devices I have seen have at least got protective boots over the spade connectors.