Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 3 of 4 1 2 3 4
Joined: Feb 2004
Posts: 14,802
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,802
Likes: 72

Blood Warmer? smile

It sounds like the pump is behaving exactly as it should, Richard. Perhaps B.Braun will incorporate an "on the fly" AIL reset button in the next version! Although I'm not so sure that would be a good idea in general, would it? Opinions welcome.


If you don't inspect ... don't expect.
Joined: Sep 2006
Posts: 745
Philosopher
Offline
Philosopher
Joined: Sep 2006
Posts: 745
We had a problem at Hereford with air in line during blood infusions using Graseby 500 pumps Link
Hopefully I've put the link in correctly.

Lee


Don't forget "we've never had it so good".
Joined: Feb 2004
Posts: 14,802
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,802
Likes: 72

As I say, I believe that the real issue is with the procedure. Shouldn't the blood be pre-warmed? Or am I missing something here?

(which could well be the case, as my own life force is a bit chilled this morning) smile


If you don't inspect ... don't expect.
Joined: May 2003
Posts: 75
Adept
OP Offline
Adept
Joined: May 2003
Posts: 75
Geoff

the ail is reset when a new infusion/therapy starts.
the use of blood warmers for every blood infusion is not practical.

Lee

thanks for the link not familar with the graseby 500 does this pump have an accumalative alarm or is it just air bubble size?

regards

richard

Joined: Sep 2006
Posts: 745
Philosopher
Offline
Philosopher
Joined: Sep 2006
Posts: 745
The Graseby giving set has three bubbles which are squeezed in sequence to give the required rate, air detection is by amount of air measured by IR in the top bubble.

Air in the circuit or bubbles produced above the pump accumulates in the top giving set bubble. The problem we had was that micro bubbles were gassing off below the pump then accumulating in the bend above the cannula.

Lee


Don't forget "we've never had it so good".
Joined: May 2003
Posts: 75
Adept
OP Offline
Adept
Joined: May 2003
Posts: 75
Hi Lee

the air will be released as the blood warms in the set, one suggestion was to let the blood warm up for approximately 10 mins before primming the line, this did not go down to well with our staff.

it takes approximately 20ml of air to kill somebody,adult that is.the accumalative air alarm is set to 3.8ml on the B-Braun pump because the pump can be used on peads.

even when blood is gravity fed the small bubbles are still there.one thing which can lead to more air in the blood is the wrough handling of the blood before being set up for infusion.

hope this is of some help!

I WOULD STILL LIKE TO HEAR FROM PEOPLE USING PUMPS OTHER THAN B-BRAUN AND HAVE AN ACCUMALATIVE AIR ALARM.I NEED TO KNOW IF YOU DO OR DON'T HAVE PROBLEMS WHILE INFUSING BLOOD.

Richard

Joined: Feb 2004
Posts: 14,802
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,802
Likes: 72

From memory I'm pretty sure that the Graseby 500 does not add up the air bubbles to give an accumulated total. As with most contemporary pumps it simply triggers an alarm (and stops pumping) when air bubbles of a certain size are detected. I don't know the size off hand, as I don't have the specs with me.

However ... our old (and much maligned) friend the Baxter Colleague does have an accumulated AIL feature! It also has the (clever) "Advance Air" feature which pumps a bit more to let you inspect the air bubble. Here is an extract from the Colleague manual:-

Code:

The air bubble setting is a configurable option.

The air sensor measures the accumulated amount of 
air detected over an amount of solution delivered. 
The amount of delivered solution depends on the 
programmed bubble size. The air alarm is triggered 
for a single air bubble greater than the set 
threshold or an accumulation of air greater than 
the threshold. The alarm threshold and accumulated 
volumes are given in the table below.

Air Bubble Setting   Accumulation Volume

  25 microlitres          0.83 mL
  50 microlitres          1.67 mL
 100 microlitres          3.33 mL
 150 microlitres          5.00 mL


It would be nice to see a chart showing the AIL characteristics of all the current pumps.

Meanwhile, see here.


If you don't inspect ... don't expect.
Joined: Feb 2004
Posts: 14,802
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,802
Likes: 72

OK ... also the Alaris GW. smile

Single bubble configurable:-

50, 100, 250, 500 microlitres (100 default)

Accumulation:-

Fixed at 500 microlitres (over 15 minutes)

Notice that (like the Colleague) the GW has a horizontal pumping mechanism. Could this be significant here, I wonder?


If you don't inspect ... don't expect.
Joined: Sep 2006
Posts: 745
Philosopher
Offline
Philosopher
Joined: Sep 2006
Posts: 745
The design of the Graseby giving set ensures that any air bubbles from above the pump accumulates in the top bubble (I tried very hard to beat this feature but failed)so it does operate as an accumulated alarm.

What it and none of the other pumps I have come across do is to measure the air close to where it goes into the patient.

The maximum amount of air we saw was 0.3mL which according to the figures I’ve seen should not be a problem. What was particularly worrying for us was that when this was occurring no one else appeared to have seen the same problem.

Lee



Don't forget "we've never had it so good".
Joined: Feb 2004
Posts: 14,802
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,802
Likes: 72

Originally Posted By: Lee S
What it and none of the other pumps I have come across do is to measure the air close to where it goes into the patient.

Hopefully we will be seeing something like this in the next generation of pumps. As I have mentioned before (but I'm still waiting for the Royalties), surely we need a miniature monitoring device (flow, pressure, bubble detection) ... plus a relief valve to release dangerous conditions as necessary ... sitting at the actual infusion site!

OK ... where's my drawing board? smile


If you don't inspect ... don't expect.
Page 3 of 4 1 2 3 4

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 4,564 guests, and 39 robots.
Key: Admin, Global Mod, Mod
Newest Members
Janet, james_on_ebme, Biomed Egypt, Nektarios K., Keisya Andretta
10,364 Registered Users
Forum Statistics
Forums26
Topics11,254
Posts74,506
Members10,364
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5