Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 2 1 2
#4373 21/02/06 11:30 AM
Joined: Feb 2005
Posts: 22
huta Offline OP
Dreamer
OP Offline
Dreamer
Joined: Feb 2005
Posts: 22
Dear sir,
A BAXTER 6201 INFUSION PUMP report F_94 when it was turned on.Who can give us some advice kindly?
Thank first!
htdocument@126.com

#4374 21/02/06 9:19 PM
Joined: Jul 2001
Posts: 235
Likes: 1
BSM Offline
Master
Offline
Master
Joined: Jul 2001
Posts: 235
Likes: 1
If I remember correctly, it is something to do with the CPU board.

Go into the service/calibration mode and clear the errors.

If it returns, you will need a new CPU board.

I will check in the manual and post again tomorrow.

#4375 22/02/06 8:58 AM
Joined: Jul 2001
Posts: 235
Likes: 1
BSM Offline
Master
Offline
Master
Joined: Jul 2001
Posts: 235
Likes: 1
F_94 - Configuration option settings checksum is not zero.

Reset config options and check battery voltages are within spec.

If the fault keeps coming back (even after the internal batteries have been replaced, the CPU board needs replacing.

#4376 24/02/06 3:21 PM
Joined: Feb 2005
Posts: 22
huta Offline OP
Dreamer
OP Offline
Dreamer
Joined: Feb 2005
Posts: 22
Dear BSM,
Thank you very much!
Would you like to tell us how to go into the service/calibration mode and clear the errors,how to reset config options and check battery voltages are within spec?
best regards,
huta

#4377 26/02/06 11:59 AM
Joined: Jul 2001
Posts: 235
Likes: 1
BSM Offline
Master
Offline
Master
Joined: Jul 2001
Posts: 235
Likes: 1
If you refer to the sevice manual, under Calibration, all the necessary instructions are there.

#4378 10/04/06 9:58 AM
Joined: Aug 2004
Posts: 88
Adept
Offline
Adept
Joined: Aug 2004
Posts: 88
I know this topic is a bit old but still very interesting to me what with more than a 100 of these units in our hospital. The pumps are under lease but still I want to know some first level repair I can do myself.

To BSM,
on the service manual that I have, the calibration starts with going into the EPROM of the unit and there is NOTHING about resetting of error messages.Please can you explain more about the one stated on your copy of the service manual?
Thanks

#4379 25/07/06 8:26 AM
Joined: Jul 2001
Posts: 235
Likes: 1
BSM Offline
Master
Offline
Master
Joined: Jul 2001
Posts: 235
Likes: 1
There is a section in the manual called: "Modifying the Configuration Option Settings" (possibly section 1.5.2).

In that section it refers to clearing the pump alarm log.

#4380 16/01/07 11:59 AM
Joined: Feb 2004
Posts: 14,807
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,807
Likes: 72
This unit was made in Singapore by Sharp, by the way.

How to clear the "Common failure F-94" error? (this is a start-up check-sum error that usually arises by having rubbish in the time and date addresses, due to all system batteries being totally depleted. The solution is to restore time and date to current, or at least valid, values):-

1) Whilst holding down "Panel lock" button at rear of unit, hold down "Stop" and “On/off" together at the front panel. This should bring up "Modify Config" in the top window.
2) Step through config options by pressing "Next" (which should now be lit) to get to "Time".
3) Enter correct time using the number keys.
4) Press green "Pri Start" key to store changed time.
5) Step to "Date", and enter correct date using the number keys.
6) Press green "Pri Start" key to store changed date.
7) Once valid entries have been made, switch off at "On/off".
8) Re-start the unit. Error code should have been cleared!

(sincere thanks to Ed Bennett, the Master of Minutiae, for helping me out with this one) smile


If you don't inspect ... don't expect.
Joined: Jan 2008
Posts: 32
mkm Offline
Visionary
Offline
Visionary
Joined: Jan 2008
Posts: 32
years later is still helpfull...

mkm #38157 05/05/09 9:53 AM
Joined: Apr 2009
Posts: 81
Adept
Offline
Adept
Joined: Apr 2009
Posts: 81
I think F94 is common fault symptom.
I have try to make this err .when you machine battery nearly empty and turn off machine. wait a few second and turn of machine.the machine sometime will generate this err.
smile



Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (KostasIatri), 1,180 guests, and 48 robots.
Key: Admin, Global Mod, Mod
Newest Members
Abdulaziz abed, Fernando Henz, Nick.A, Tdayve, AdrianT82
10,373 Registered Users
Forum Statistics
Forums26
Topics11,262
Posts74,527
Members10,373
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5