can somebody help me with a service manual or with error code for an Caleo Incubator? I don't know the exact model but i'll find it today( it's not so new )- i get the following error after power on:

Device Error 40

The red led is ON and also the Yellow LED at AIR stay's ON.
( my email is onkynm@yahoo.co.uk )

Last edited by Onky; 31/05/10 2:10 PM.
Joined: Apr 2002
Posts: 135
Expert
Offline
Expert
Joined: Apr 2002
Posts: 135
I have seen this error when there has been a problem with a skin temperature probe.
After the probe has been removed and the Caleo "rebooted" the alarm does not occur again.
I have also seen this error when we perform an overtemperature test, again after switching off the the Caleo, the error is cleared when powered up again.
View your error logs for other clues and check all the cables and connections.


New beginning. New location. New adventure. Old grumpy me!
Joined: Feb 2009
Posts: 1,908
Likes: 18
Hero
Offline
Hero
Joined: Feb 2009
Posts: 1,908
Likes: 18
Onky, I have emailed you the service manual for the Caleo, this is for the new version, hope it contains the information that you require.


I am not Flippant, I am Smart
Joined: Oct 2007
Posts: 75
Adept
Offline
Adept
Joined: Oct 2007
Posts: 75
Error 40 have seen it in error logs, was advised by drager, if no reoccurance to ignore. manual says its a variables conflic, cause software problem or external interface. Remsdy, switch device off and on again if this does not help proceed as follows. replace WT2 controller PCB.

has paul advised i would remove all external leads and devises and reseat all internal connections, long process but hay ho.

Paul doolan at drager is also helpfull


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 41,138 guests, and 14 robots.
Key: Admin, Global Mod, Mod
Newest Members
Abdulaziz abed, Fernando Henz, Nick.A, Tdayve, AdrianT82
10,373 Registered Users
Forum Statistics
Forums26
Topics11,263
Posts74,531
Members10,373
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5