|
|
Joined: Jun 2010
Posts: 63
Scholar
|
OP
Scholar
Joined: Jun 2010
Posts: 63 |
hi everybody i have the problem with stryker 30Lhigh flow insufullator, problem is m/c shows E04 at Gasflow any one service manual,pls send me or guide me to rectify the problem,advance thanks
|
|
|
|
|
Joined: Dec 2009
Posts: 121
Savant
|
Savant
Joined: Dec 2009
Posts: 121 |
I've got the service manual for the 40L and it states to check the gas supply. Hope that helps
|
|
|
|
|
Joined: Jun 2010
Posts: 63
Scholar
|
OP
Scholar
Joined: Jun 2010
Posts: 63 |
thanks sir,my email id : rajan_bme@yahoo.co.in pls send to me
|
|
|
|
|
Joined: Dec 2009
Posts: 176 Likes: 1
Mentor
|
Mentor
Joined: Dec 2009
Posts: 176 Likes: 1 |
Hi NickM. Do yu can send me service manual, too? hugo_68@yahoo.com Do you have other Stryker service manuals? Lighth sources per example. Thanks
|
|
|
|
|
Joined: Dec 2009
Posts: 121
Savant
|
Savant
Joined: Dec 2009
Posts: 121 |
Hi Guys, I work for Stryker as a Field Service Engineer, unforunately for you guys all the manuals I have are in electronic form in a program we use, I could probably get extracts if required, but if you post any problems you have I will try to help or atleast point you in the right direction.
|
|
|
|
|
Joined: Jun 2009
Posts: 796 Likes: 13
Philosopher
|
Philosopher
Joined: Jun 2009
Posts: 796 Likes: 13 |
Hi Nick, how's the new job going?, PM me if you like (he used to be one of my guys for those who are going....."what the hell?")
Dave
Last edited by DaveC in Oz; 09/08/10 11:29 AM.
Thoughts and information provided on this forum are mine and mine alone and do not necessarily reflect the policy of NSW Health. They may also be complete bollocks!!
|
|
|
|
|
Joined: Jun 2010
Posts: 63
Scholar
|
OP
Scholar
Joined: Jun 2010
Posts: 63 |
When put on the m/c it shows error 4, kindly guide me,advance thanks
|
|
|
|
|
Joined: Dec 2009
Posts: 121
Savant
|
Savant
Joined: Dec 2009
Posts: 121 |
Like I stated in a previous post you need to check the gas supply, its possibly the internal regulator requires calibration and or repair/replacement.
"4.6.1 Checking Gas Supply Connection/Gas Inlet Filter (See Fig. 4-6-1-1) The gas inlet filter (2) may be plugged or soiled due to Soiling unclean gas or gas supply residues. This type of soiling, however, cannot be detected externally. Heavy soiling is especially noticeable during higher flow rates and results in the actual gas flow falling short of the nominal gas flow during active insufflation. If cleaning the surface of the gas inlet filter does not yield any improvement, the gas supply connection (1) has to be replaced completely"
Keep in mind this is for the 40L
|
|
|
|
|
Joined: Jun 2010
Posts: 63
Scholar
|
OP
Scholar
Joined: Jun 2010
Posts: 63 |
|
|
|
|
|
Joined: Dec 2009
Posts: 121
Savant
|
Savant
Joined: Dec 2009
Posts: 121 |
Let me know if you need the part numbers for a replacement regulator or gas inlet filter assy
|
|
|
|
1 members (daisizhou),
1,727
guests, and
39
robots. |
|
Key:
Admin,
Global Mod,
Mod
|
|
|
Forums26
Topics11,278
Posts74,590
Members10,391
| |
Most Online69,877 Jul 17th, 2026
|
|
|
|