Dear all
I have a Criticare 508 Patient Monitor.I have calibrated the nibp pressure with mmHg meter. The problem is that the initial inflation pressure is very low that is 110mmHg. So in its first attempt it fail to get reading, then it inflate again goes to 125mmHg & again no reading same way it gets a reading after 3 or 4 attempts. the patients are not satisfied with it, they need to get reading in first attempt. Can any one share their experience if i can change any setting in setup so that the inflation pressure should start from say 160 or 170 mmHg.
How to get into its setup menu.
Hope to get quick reply.
Thanks in advance & regards
LAIQ

Joined: Jun 2009
Posts: 796
Likes: 13
Philosopher
Offline
Philosopher
Joined: Jun 2009
Posts: 796
Likes: 13
Sounds like the monitor is in neonatal mode (or paediatric) hence the low starting pressure. Don't know how to access the set ups however but perhaps someone else can assist with that,


Thoughts and information provided on this forum are mine and mine alone and do not necessarily reflect the policy of NSW Health. They may also be complete bollocks!!
Joined: Apr 2005
Posts: 380
Laiq Offline OP
Sage
OP Offline
Sage
Joined: Apr 2005
Posts: 380
Thanks Davec
Yes i am thinking on same line. There is a CONFIG button by this u can change some settings, in there Spo2 mode showing ADULT, dont know about NIBP settings as there is none.
Thanks for your help anyway. Hope some one can get my answer.
Regards

Joined: May 2002
Posts: 90
Adept
Offline
Adept
Joined: May 2002
Posts: 90
I think the Criticare NIBP takes its readings as the cuff inflates and not while deflating like Dinamap etc. It is claimed this is more comfortable for the patient.

Joined: Apr 2005
Posts: 380
Laiq Offline OP
Sage
OP Offline
Sage
Joined: Apr 2005
Posts: 380
Dear Sir
Can u explain a bit why it is getting low inflation pressure during first two attempts. I have replaced the cuff & pipes with new one, but still same behaviour. Is there any operating manual for it so that i can get the picture of what may be the problem.
Waiting for quick reply.
Regards
LAIQ


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
0 members (), 7,223 guests, and 92 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,288
Posts74,594
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5