|
|
Joined: Jan 2003
Posts: 39
Visionary
|
OP
Visionary
Joined: Jan 2003
Posts: 39 |
Hi fellow Biomeds
Can anyone help with regards to methods for checking APC Medical related external pacemakers?
I have been looking into testing everything that they ask you to do, using a more dedicated Pacemaker Analyzer. I have here the Fluke SigmaPace 1000 Analyser.
I have managed to do most tests, but find it difficult to get the Manufacturer's spec for SRP (Sensed Refractory Period) of 150ms. I get a value of around 237ms using the Fluke analyser.
I found it quite difficult to translate what APC want you to do, using the pacemaker analyser. Any others have this issue?
Any external pacemaker geeks out there with in hints or resources to get my teeth into the operation and testing of these?
All help and advice is appreciated.
The chief function of your body is to carry your brain around.
|
|
|
|
|
Joined: Jul 2002
Posts: 2,020
Hero
|
Hero
Joined: Jul 2002
Posts: 2,020 |
I have not used the Fluke tester but in the past with other testers the output waveforms had to be the correct shape for the tester to read this time correctly. If you do not have a square wave the tester has to measure the time from a rising voltage, where exactly does it take the start and stop times (at what voltage) for the timer? Either the tester is expecting a square wave or is making a calculation (estimation) based on the rise time of the waveform.
You have to check the settings on the tester and pacemaker are correct.
Robert
My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
|
|
|
|
|
Joined: Feb 2004
Posts: 14,807 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,807 Likes: 72 |
The SigmaPace certainly seems comphrensive enough. I would love to have a play!  Meanwhile I'm wondering what APC recommends when it comes to testing this kit? What test equipment do they suggest, and all the rest?
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Jan 2003
Posts: 39
Visionary
|
OP
Visionary
Joined: Jan 2003
Posts: 39 |
Cheers RoJo, that helps in understanding what the test entails.
Hi Geoff, APC Medical recommend using a function generator, oscilloscope, a counter, and a delay unit used to trigger the function generator in ms.
We just don't have that type of kit around anymore that is calibrated regularly. Hence, the query in relation to using the Pacemaker Analyzer.
The chief function of your body is to carry your brain around.
|
|
|
|
|
Joined: Oct 2003
Posts: 51
Scholar
|
Scholar
Joined: Oct 2003
Posts: 51 |
For the reasions that Robert describes most pacemker testers will in effect only give you an indication of function.
If you want to be accurate then the only real method is from 1st principles and some good basic electronic test equipment. Therefore in no order of priority, normally the following is required:
Manufacturers test proc sheet with test limits - this will normally give you the method, test equipment settings, and the range of acceptable results. Oscilloscope Multimeter Suitable Test Body / test load Function Generator for sq and triangular wave signals - must have variable amplitude and frequancy abilities Counter timer with external trigger abilities Power Supply.
Shed loads of patiance can also be useful.
|
|
|
|
|
Joined: Feb 2004
Posts: 14,807 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,807 Likes: 72 |
Sounds like an annual return for cal by APC might be a sensible option here. Either that or sending off to some third-party or other able to check these things out properly. If, indeed, such a third-party exists! Yet another area, perhaps, where a specialist referral workshop within the NHS (that is, one with all the test kit) might also be a good idea! Any volunteers? 
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Oct 2003
Posts: 51
Scholar
|
Scholar
Joined: Oct 2003
Posts: 51 |
Yee Gods, I am a bit worried that we as a profession don't have what I would regard as the essential basics around any more. The fact that we seem to have to moved a reliance an specfic "test kit" which is actually only functional inspection and verification kit is quite scarey really. May be I'm just getting too old.
In relation to these devices yeah the list of equipment does seem a bit excessive and the process a tad anal but it does work, really don't take that long to do; is well with the abilities of any engineering prectitioner with a good general electronics background and is really the only true way of adjusting these types of devices.
However I do agree that you can adjust these devices using units like the sigma pace or the pma or others, and indeed we do here, but you need to know exactly what you are looking at when read the displays of these devices. Mine are told that devices such as Sigma pace - which is very nice indeed are for indication only- if they have an doubt about what they are seeing then back to Mr. 1st Principles. That by the way goes for everything not just these devices.
After all Geoff is this not the true art of "biomedism" that we should all be practising? Because if we don't and we are placing all our reliance on automated or semiautomated "testing and inspection" equipment where we don't fully understand what its doing and why its doing it then we are on a very sticky slope indeed.
|
|
|
|
|
Joined: Feb 2004
Posts: 14,807 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,807 Likes: 72 |
Indeed, Paul. Back to Basics (along with its Mate, Thinking Simple) always gets my vote! 
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Jul 2002
Posts: 2,020
Hero
|
Hero
Joined: Jul 2002
Posts: 2,020 |
And you can use the scope, function generator etc. to test other medical kit rather than having to fork out for more specialist test equipment.
And I am sure the calibration certificates for this kit would be far less than the manufacturer/distributer of the medical test kit would charge as they have the monopoly.
Robert
My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
|
|
|
|
|
Joined: Jan 2003
Posts: 39
Visionary
|
OP
Visionary
Joined: Jan 2003
Posts: 39 |
I have to agree with the above statements in regards to competency in 'biomed' skills, but, these analysers are on the market and we did buy one. I would therefore like to get the full functionality out of it.
It has been a bit of a relearning curve for me this week in regards to thinking about first principles, using a custom made piece of test equipment. I have also found myself writing a test protocol to accomodate manufacturer's specification, but in light of using our Fluke Analyser.
I have sent a pacemaker back to APC to do, as they only charge £100 for the privilege. I want to see a full service report with documented results so I can replicate them using the Analyser.
The chief function of your body is to carry your brain around.
|
|
|
|
2 members (Daniel Falcon, Huw),
1,150
guests, and
39
robots. |
|
Key:
Admin,
Global Mod,
Mod
|
|
|
Forums26
Topics11,261
Posts74,526
Members10,371
| |
Most Online59,530 Apr 30th, 2026
|
|
|
|