Hello

I need to get a new Infusion Pump Tester, but which one would you recommended?

Hervé

#15037 13/07/05 1:50 PM
Joined: Jul 2000
Posts: 969
Likes: 1
Roy Offline
Philosopher
Offline
Philosopher
Joined: Jul 2000
Posts: 969
Likes: 1
They all seem to have problems and as we use Alaris pumps we follow their test procedure and use mechanical guages to measure the plunger displacement and force.

We've used the Ultramedic analyser in the past with the intention of saving time - but in fact we found that we were spending a lot of time chasing false failures caused by air bubbles, sticking syringes, eroneous readings etc.

The old fashioned way seems to be the best - even if that means purchasing an electronic stopwatch to alarm when you need to measure the fluid in the burrette !


Today is the day you worried about yesterday - and all is well !
#15038 18/07/05 2:45 PM
Joined: Oct 2004
Posts: 23
Dreamer
Offline
Dreamer
Joined: Oct 2004
Posts: 23
I find the metron lagu very good but do your own research to make sure it is suitable for you.

http://www.metron-biomed.com/products.php?m_name=lagu&lang=Eng

Denis.


The art of medicine consists of amusing the patient while nature cures the disease.
#15039 18/07/05 9:51 PM
Joined: Jul 2005
Posts: 1
Newbie
Offline
Newbie
Joined: Jul 2005
Posts: 1
Hi guys,
I've tried a few testers too including the IDA and SPC-1 - which one are you refering to Roy?
I have to say the IDA has never caused me any problems as long as you prime it right using the old pre-in-line syringe method - the SPC-1 is pretty good but can be complicated in my opinion and has a habit of giving odd readings with some worm driven infusion pumps.
Indeed sometimes its always better to go back to basics and use raw tools but give the IDA1/2 a try. Hope this helps.
Cheers
Andrew smile


"When it's raining - why walk faster...it is raining just as hard up ahead!"
#15040 19/07/05 8:45 AM
Joined: Apr 2002
Posts: 188
DAS Offline
Mentor
Offline
Mentor
Joined: Apr 2002
Posts: 188
Quote:
Originally posted by szymczak:
Hello

I need to get a new Infusion Pump Tester, but which one would you recommended?

Hervé
We use the IDA 4+, but only for volumetric pumps. With correct use you should have no problems. For syringe drivers I have previously used Ultramedics syringe driver tester Ultramedic
In both cases these devices are for checking & verification. For calibration & setup you need to follow the manufacturers recommendations.


Never under-estimate the predictability of stupidity
#15041 26/07/05 2:01 PM
Joined: Mar 2003
Posts: 54
Scholar
Offline
Scholar
Joined: Mar 2003
Posts: 54
i am using metron lagu in our hospital. Do not like its noise very much. And limited function and data can be achieved.

I used IDA4 plus in another hospital. The Hydrograph Graphics Software is very good for testing pumps' stabilization.

Joined: Jul 2010
Posts: 84
Adept
Offline
Adept
Joined: Jul 2010
Posts: 84
Could someone please guide me from where i can download the same software??

Joined: Jul 2002
Posts: 2,020
Hero
Offline
Hero
Joined: Jul 2002
Posts: 2,020
I think you will find that people are talking about hardware for testing - or do you mean software for a perticular tester?


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
Joined: Jul 2010
Posts: 84
Adept
Offline
Adept
Joined: Jul 2010
Posts: 84
RoJo sir,
I am aware of that the discussion is regarding the hardware available to test Infusion Devices...
I have 2 IDA 4+ devices with me... but with out the software the efficacy of the unit is rather low...

Joined: Feb 2003
Posts: 362
Likes: 3
Sage
Offline
Sage
Joined: Feb 2003
Posts: 362
Likes: 3

The Hydrograph IDA4 + copyrighted software is available from your Fluke Biomedical service agent, it is not freeware, or freely available in the public domain.

The IDA4 + works fine without Hydrograph, I would dispute your assertion that "with out the software the efficacy of the unit is rather low"

Some things are free, and some things you have to pay for, (that's just the way it is).

Page 1 of 3 1 2 3

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (WAYNE MILSOM), 9,533 guests, and 22 robots.
Key: Admin, Global Mod, Mod
Newest Members
Chris 11, j9_PLC, nece, Vitya, Shenzhen007
10,358 Registered Users
Forum Statistics
Forums26
Topics11,249
Posts74,483
Members10,358
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5