Dear All I just want your expert opinion, can i convert 110V MR730 Fischer & Paykel Humidifier to 220V input by replacing the transformer or i have to do more changes. Please guide me if u can. Thanks in advance & Regards. LAIQ
|
|
|
|
|
Joined: Oct 2010
Posts: 12
Novice
|
Novice
Joined: Oct 2010
Posts: 12 |
110 Volts F&P Humidifiers can not be changed to 220 volts unless F&P supply you with the internal transformer of 220 volts which I really doubt. I have never seen a part number for the replacement of a 220 volt transformer in the part's list.
Try to contact F&P and ask. I personally, under any circumistances will not allow the use of a step-up transformer to any medical equipment that is connected directly to patients.
If your hospital policy does allow it, then you can attach an Isolated step-up transformer to the base of the unit's trolley provided it will not tip over.
Good Luck to you
|
|
|
|
|
Joined: Feb 2004
Posts: 14,798 Likes: 71
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,798 Likes: 71 |
Nonsense! Of course they can be modified. Fisher & Paykel are not the only suppliers of transformers in the world!  @Laiq: have you got it done yet?
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Feb 2004
Posts: 14,798 Likes: 71
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,798 Likes: 71 |
Check out the manual!  Pity that the circuit diagrams are missing though. Perhaps someone else can do better!
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Aug 2001
Posts: 797 Likes: 1
Philosopher
|
Philosopher
Joined: Aug 2001
Posts: 797 Likes: 1 |
For all the hassle and time it will take it maybe easier to look at buying a used operational one. Or even see if someone has a broken one you can have.
|
|
|
|
|
Joined: Jul 2009
Posts: 2
Newbie
|
Newbie
Joined: Jul 2009
Posts: 2 |
The Heator plate is also powered from AC mains. So you may need to change both. Heater plate Assembly 150W 230V P/N 043040695 and Transformer Assembly (MR700 series) 230V (double insulated) P/N 043041287
Joy
|
|
|
|
|
Joined: Apr 2005
Posts: 380
Sage
|
OP
Sage
Joined: Apr 2005
Posts: 380 |
Dear Geoff Thanks bro, i have checked thoroughly (without circuit diagram) Joy is right heater is 110V also, so impossible to modify it. So currently i am stuck, whole lot is useless for time being.. :-( Last option is to buy faulty 220V pcs or broken body pcs, then replace heaters & tansformers. Thanks to all for ur feedback.
|
|
|
|
|
Joined: Feb 2004
Posts: 14,798 Likes: 71
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,798 Likes: 71 |
Yeah, sorry about that. The only other thing I can think of is to look out for a box of them at the auctions!  But (to be honest) they don't seem to turn up very often.
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Apr 2005
Posts: 380
Sage
|
OP
Sage
Joined: Apr 2005
Posts: 380 |
Dear Geoff It was my mistake, i should not have bought USA made model, it should be UK made due to input voltage & my country supply voltage.. Dont worry i have many other uses, if i told u here so many will be marketting their products.. Thanks for your concern any way.. May ALLAH give u reward for this. Thanks freind & brother for helping many like us in this forum. Lots of prayers for u.
|
|
|
|