My Dear "COLLEAGUES",
I won't hear a thing said against them, as another real colleague said to me, in the tone of the immortal words of Sir Winston Churchill - Never in the field of NHS conflict, have so many people had something worked so far up them, by so few.
In reply to Graham, Member # 548 and all other interested fellow Tech's.
We have two test methods for the Baxter Colleague, the first, which is basically the preferred Baxter method, is a specially purchased Avery Berkel, model TA3001A scale, with a resolution of 0.1g and a tolerance of + or – 0.3g. The pump discharges into a 500 ml Pyrex flask.
This scale is connected to a dedicated Test PC, running an in-house Access developed system on Windows 95. The display is along the lines of the IDA4+ but to our requirements and provides; average rate, % error, volumetric error and many other customised details for record storage purposes and is linked via our Hospital Intranet to our EBME Data Management System.
The scale is calibrated in-situ, annually, by Avery Berkel, the results are excellent, providing a graphic record plus a Tabular one, to an accuracy demanded by Baxter, without the hassle of stop watches or standing over it waiting for results.
Having said the above we also have the, not to be disparaged, IDA4+, which gives excellent results also, to an accuracy very compatible with our own system.
Why develop our own you may say, well Baxter can't admit to the accuracy of all the IDA systems in use, for the very simple reason that a large percentage of users don't have them calibrated by Ultramedic on an annual basis, and also their "housekeeping' methods are probably dubious. They need to be used ONLY with non-saline based solutions, demineralised water or distilled water for the best results.
After use they should be flushed out and preferably blown dry with a small air pump to clear the system of all water deposits which can leave marks on the micro pipette measurement system. Failure to do this will result in errors building up. We can say this with all the confidence in the world – our head of department previously serviced and calibrated them at Ultramedic, and knows all about neglect by users. Nurses can't have all the credit for bad custom & practice.
Baxter say in their; "Colleague Pump – Volume Delivery Accuracy Test Document', dated November 2001 and I quote, “ As with all measuring equipment it is important to ensure routine maintenance and calibration is carried out to ensure the tester remains with in its operating accuracy”.
To reply to Chris-H, Member 176, yes we are seeing a lot of error codes, mainly caused by the darling nurses doing things wrong and causing lock-outs and tube misloaded faults causing the usual manual tube release messages and lock-up problems. Also using the pump until the battery is exhausted.
Though the latest stated failure message, purported to be in the LED display, is ITP, any clues any one, as to what ITP refers, the Pump History record shows nothing? Other wise accuracy has been in the last year our main problem, although with the new more accurate giving sets now in use from Baxter this hopefully appears to be a thing of the past.

Ian-C