Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 5 1 2 3 4 5
#49797 21/10/10 5:35 PM
Joined: Jul 2010
Posts: 16
Novice
OP Offline
Novice
Joined: Jul 2010
Posts: 16
Hello everyone
I have a problem with the LOGIQ 500 cpu card (12). When switched on, the system loads the first two notches colored below the word LOGIQ, and shows the number 2135131-2 (error number?). I want to try to repair the card. It features a Motorola MVME147-010 main board (16 MHz, 4MB RAM, 1PIO, SCSI).
I removed the battery of the timekeeper to measure it, and it is low, I think the same mother board has lost the BIOS settings. Can anyone help me? Thanks in advance

Joined: Jul 2010
Posts: 16
Novice
OP Offline
Novice
Joined: Jul 2010
Posts: 16
can anyone help me?

Joined: Jul 2010
Posts: 16
Novice
OP Offline
Novice
Joined: Jul 2010
Posts: 16
hey... my friends... Can anyone know the error 2135131-2 ? Please help...

Joined: Feb 2004
Posts: 14,806
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,806
Likes: 72

Sorry to hear that no-one has been able to help so far. smile

I doubt that is an error number. My guess would be that it is the unit's serial number, or some such thing. Either way, it sounds like you have either lost some basic settings, or have a corrupted hard drive or something like that.

What else is going on ... is the unit booting up?

Anyway, you may need some expert assistance. See your PM's.


If you don't inspect ... don't expect.
Joined: Feb 2009
Posts: 1,908
Likes: 18
Hero
Offline
Hero
Joined: Feb 2009
Posts: 1,908
Likes: 18
The main problem that we have had with GE Ultrasound machines, is that when we have had a software problem and the unit is down, once any of the boards has been removed and then returned we get the message that the license needs to be renewed/upgraded. I think it is GE's way of stopping un-authorised service of their machines.


I am not Flippant, I am Smart
Joined: Feb 2004
Posts: 14,806
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,806
Likes: 72

Yes, you've got to safe-guard those "key codes" (strings, or whatever it is they call them).

It's the same sort of deal with most of the ultrasound units. We can only but shudder to imagine what matters will be like when such kit is "no longer supported" in a few years time. Just about the same time, of course, as they will be turning up on the pre-owned and/or donated kit scene!

But hey, who knows ... perhaps at that time the gougers will do the decent thing and release all the info into the Public Domain! Dream on. frown


If you don't inspect ... don't expect.
Joined: Jul 2010
Posts: 16
Novice
OP Offline
Novice
Joined: Jul 2010
Posts: 16
Dear Geoff and Neil,
thanks for the reply.
In fact now I'm sure I was right.
I was able to connect via remote, the machine also confirmed:

VENUS *** SYSTEM RESET / START ***
*** *** BBRAM contents are corrupted
VTSH>
VTSH> *** VENUS *** SYSTEM READY
PWRD: rt_reqprdmsg error [1100th]

BBRAM : Battery Backup Ram

Now I just have to figure out how to enter the system ID of the machine. Have you by chance P9030TB Logiq 500 proprietary manual?

Joined: Jul 2010
Posts: 16
Novice
OP Offline
Novice
Joined: Jul 2010
Posts: 16
Originally Posted By: Neil Porter
The main problem that we have had with GE Ultrasound machines, is that when we have had a software problem and the unit is down, once any of the boards has been removed and then returned we get the message that the license needs to be renewed/upgraded. I think it is GE's way of stopping un-authorised service of their machines.



Hello Neil, maybe this happens on newer machines. The Logiq 500 CPU boards and I have always changed with no problem.

Joined: Jul 2010
Posts: 16
Novice
OP Offline
Novice
Joined: Jul 2010
Posts: 16
Originally Posted By: Geoff Hannis

Sorry to hear that no-one has been able to help so far. smile

I doubt that is an error number. My guess would be that it is the unit's serial number, or some such thing. Either way, it sounds like you have either lost some basic settings, or have a corrupted hard drive or something like that.

What else is going on ... is the unit booting up?

Anyway, you may need some expert assistance. See your PM's.


You are right Neil, the machine has lost the data on BBRAM. This is the message I get through the terminal:

*** VENUS SYSTEM RESET/START ***
*** BBRAM contents are corrupted ***
VTSH>

Has anyone had the same problem?

Joined: Jul 2010
Posts: 16
Novice
OP Offline
Novice
Joined: Jul 2010
Posts: 16
Dear Friends,
we have resolved the problem. It's very simple. GE ask 8.000 euro for the PCB. 3 hour of my work, and 6.75 euro for the battery from farnell et voilà, the CPU work. How strange this world... Saluti !

Joined: Feb 2004
Posts: 14,806
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,806
Likes: 72

Nice one, Alessandro ... that's just the sort of thing we like to hear!

Bravo! smile


If you don't inspect ... don't expect.
Joined: Jul 2010
Posts: 16
Novice
OP Offline
Novice
Joined: Jul 2010
Posts: 16
I'm available for any friend with the same problem, thank you!

Joined: Nov 2010
Posts: 1
Newbie
Offline
Newbie
Joined: Nov 2010
Posts: 1
Hi my friend,

I'm having the same problem, but on Logic 400. Could you help me?

Thank you.

Joined: Jul 2010
Posts: 16
Novice
OP Offline
Novice
Joined: Jul 2010
Posts: 16
Hello Dimitri, sorry for the delay.
Write me a private message with your email, we'll see what we can do.

Joined: Nov 2008
Posts: 22
Dreamer
Offline
Dreamer
Joined: Nov 2008
Posts: 22
so alesandro,
Did you finally change the pcb board

Happy Xmas

DJIELA

Joined: Jul 2011
Posts: 5
Newbie
Offline
Newbie
Joined: Jul 2011
Posts: 5
I have a logiq 500md whith this problem. could you help me?

Joined: Sep 2011
Posts: 2
Newbie
Offline
Newbie
Joined: Sep 2011
Posts: 2
Hello friends,

I'm having the same problem, but with CPU on Logiq 400. It just can not start after changing the battery of the timekeeper. Could someone help me?

Thank you very much!

Joined: Feb 2004
Posts: 14,806
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,806
Likes: 72

Welcome to the forum. smile

According to the Logiq 400 manual ... after replacing the timekeeper battery you need to enter a System ID into the timekeeper RAM (and perhaps set up other default settings). frown

See page 4-31 of the manual.


If you don't inspect ... don't expect.
Joined: Sep 2011
Posts: 2
Newbie
Offline
Newbie
Joined: Sep 2011
Posts: 2
Unfortunately, I can't use Utility Menu to enter a System ID, because the Logiq 400 hangs during starting selftest.

Joined: Feb 2004
Posts: 14,806
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,806
Likes: 72

Sorry to hear that. frown

Check your messages.


If you don't inspect ... don't expect.
Joined: Apr 2012
Posts: 1
Newbie
Offline
Newbie
Joined: Apr 2012
Posts: 1
I'm having a problem with a ge logic 400. It wouldn't boot up so I replaced the timekeeper battery. Now I need to reprogram the date into ram but have been unable to do this. Can someone tell me how to do this?

Joined: Feb 2003
Posts: 362
Likes: 3
Sage
Offline
Sage
Joined: Feb 2003
Posts: 362
Likes: 3
The full service manual is freely available is the GE Healthcare Common Documentation Library.
But in essence, you need to:

Top Menu Select
Highlight Set Up
Sub menu to User Utility
Select Time Adjustment
Input new time by using keyboard the press enter
Input format for time is HH:MM
Put a "-" character in front of the time to decrease the current time

Joined: Feb 2003
Posts: 362
Likes: 3
Sage
Offline
Sage
Joined: Feb 2003
Posts: 362
Likes: 3
Probably worth noting that you can only adjust the time within plus or minus 12 hours.
Service Manual does not specifically identify how to alter date!
This may be taken from system BIOS?

Joined: May 2012
Posts: 3
Newbie
Offline
Newbie
Joined: May 2012
Posts: 3
Hi friends,
Also I'm repairing a logic 500 with the same problem. I replaced the battery timekeeper, was discharged 0.5V instead of 2.8V but will not start the sw. Only writes: CPU MEMORY TEST: SUCCESS and stop 2135131-4. I can not enter the system ID, no display lights on the keyboard.
What can be? Thanks in advance
Gerardo

Last edited by geros79; 10/05/12 9:28 AM.
Joined: May 2012
Posts: 3
Newbie
Offline
Newbie
Joined: May 2012
Posts: 3
Hi friends,
i connect Logiq 500 to laptop with hiperterminal. Now i can read this message:
*** VENUS SYSTEM RESET/START ***
*** BBRAM contents are corrupted ***
VTSH>
VTSH>*** VENUS SYSTEM READY ***
PWRD : rt_reqprdmsg error [1100a]
Can you help me?

Joined: Mar 2013
Posts: 2
Newbie
Offline
Newbie
Joined: Mar 2013
Posts: 2
hi
Alessandro can i have please your e-mail my own is dedieu_03@yahoo.fr
GOD Bless


Joshua 1:1-8-9
Joined: Mar 2013
Posts: 2
Newbie
Offline
Newbie
Joined: Mar 2013
Posts: 2
Hi
Alessandro please can you e-mail me at dedieu_03@yahoo.fr i have a probleme with my Logiq 500 Pro the same like you


Joshua 1:1-8-9
Joined: Sep 2013
Posts: 10
Novice
Offline
Novice
Joined: Sep 2013
Posts: 10
Dear dealers,
we can supply logiq 400/500 hard disks
Time keeper replacement is not successfull to solve booting isuues of logiq 400/ 500 machines.
we will repair Cpu ethernet address PCb of Logiq 400/ 500 machines
which is having problem like cpu memmory test success, no further bootting problems

shyjumon
www.ultrasounds engineers.com

shyjumon@hotmail.com
india

Joined: Apr 2009
Posts: 25
Dreamer
Offline
Dreamer
Joined: Apr 2009
Posts: 25
Hello
Dear Colleagues,
Write me on private e-mail address f.panci@panci-electronic.com if you need help regarding with Timekeeper and booting problem on GE Logiq 400/500. I can help everybody for free.
Best regards.


With best regards,
Fredi PANCI,
Biomedical Engineer.
Joined: Jan 2014
Posts: 1
Newbie
Offline
Newbie
Joined: Jan 2014
Posts: 1
Frepa_AL,
Thank you a lot for a quick response and invaluable help, that you gave! The instruction you send me was very helpful. The device was put into operation.

With Best Regards! Kirill.

Last edited by Kirill; 27/01/14 5:38 PM.
Joined: Apr 2009
Posts: 25
Dreamer
Offline
Dreamer
Joined: Apr 2009
Posts: 25
Everybody is welcome,
Thank you Krill, have good luck. . . .


With best regards,
Fredi PANCI,
Biomedical Engineer.
Joined: Jul 2005
Posts: 601
Philosopher
Offline
Philosopher
Joined: Jul 2005
Posts: 601
Is this the most widely discussed topic? I lost count of how many different countries the contributors have come from.
Truly a global post.

Joined: Jul 2015
Posts: 2
Newbie
Offline
Newbie
Joined: Jul 2015
Posts: 2
Hi friends, I was searching information about how can i fix the timekeeper problem, and found this forum I hope you will can help me.
greetings from mexico and thanks for your time.

Joined: Jun 2010
Posts: 63
Scholar
Offline
Scholar
Joined: Jun 2010
Posts: 63
send your email id

Joined: Apr 2009
Posts: 25
Dreamer
Offline
Dreamer
Joined: Apr 2009
Posts: 25
I can help you on this issue.
Write me back if you are still need this solution.


With best regards,
Fredi PANCI,
Biomedical Engineer.
Joined: Sep 2005
Posts: 108
Savant
Offline
Savant
Joined: Sep 2005
Posts: 108
Hi,
I'm searching information about how can fix the timekeeper problem - LOGIQ 400
ipetev@yahoo.com

Thank you.

Joined: Sep 2005
Posts: 108
Savant
Offline
Savant
Joined: Sep 2005
Posts: 108
I found the necessary information about how can fix the timekeeper problem - LOGIQ 400
IPetev

Joined: Sep 2010
Posts: 412
Likes: 2
Sage
Offline
Sage
Joined: Sep 2010
Posts: 412
Likes: 2
Nice one ipetev..share the links.


--- PATIENT SAFETY FIRST ---
Joined: Feb 2017
Posts: 1
Newbie
Offline
Newbie
Joined: Feb 2017
Posts: 1
Hello Fredi
Could you please share information about how to 1) Install the system ID number and then set up two items, Video Signal and 2) Power Frequency among four system preset parameters.
the former it suppose is easy, but the second, needs the proprietary manual instruction in the P9030TB, which i do not have.
An finally Just for curios I would like to know, if it is necessary to made a User data Back up, before replace de timekeeper RAM.

regards

Joined: Apr 2016
Posts: 1
Newbie
Offline
Newbie
Joined: Apr 2016
Posts: 1
Hello
I have problem with logic 500 pro series when i could not change the language to English. I have tried many times but still in French .
I Want to measure the timekeeper put I'm afraid if the software damage .
Please if someone can help me contact with me in Hotmail

hael-almesbahi@hotmail.com

Joined: Apr 2023
Posts: 1
Newbie
Offline
Newbie
Joined: Apr 2023
Posts: 1
Hello guys,
Im having the same problem as most of you here just hope its not too late and that someone who knows is still here smile
So its Logiq 500 and its stops booting on second bar with same numbers showing. I got the replacment battery for the timekeeping but it still stops at second bar when bootin up.
Anyone who remembers what the fix was?
Thanks in advance

Joined: Feb 2007
Posts: 302
Likes: 16
Master
Offline
Master
Joined: Feb 2007
Posts: 302
Likes: 16

Page 1 of 5 1 2 3 4 5

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (BiomedR), 3,061 guests, and 10 robots.
Key: Admin, Global Mod, Mod
Newest Members
Tekie 77, Szymon1990, Janet, james_on_ebme, Biomed Egypt
10,366 Registered Users
Forum Statistics
Forums26
Topics11,256
Posts74,519
Members10,366
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5