we are currently looking at replacing our old IVAC572 volumetric pumps in our Peads unit, I have a couple of questions - does anyone use Fresenius Agilia devices (Good or Bad ?), and then what pressure settings are used for occlusions - I am having a hard time getting the users away from the 50mmHg settings as used with the 572. All replies / info gratefully rx'd. Ta.

Joined: Aug 2007
Posts: 306
Master
Offline
Master
Joined: Aug 2007
Posts: 306
Our lot uses the 50mmHg philosphy until I pointed out that it all depends on the height of the patient to pump. Horrified was the expression when it showed greater than 50mmHg without alarm.

Does the Agilia have a pressure transducer that can be altered or is it fixed.

This as we know is hard to get them away from the brainwashing 50mmHg as we are still trying aswell. 75mmHg was too high!!

Is it just infusions you are trying for or is it syringe drivers aswell.

Joined: Jul 2002
Posts: 2,020
Hero
Offline
Hero
Joined: Jul 2002
Posts: 2,020
David,
Have you already got the Fresenius pump? BBraun have new software with predictive alarms that looks interesting.
We are in the same boat want to replace the 572 but there is no obvious replacement.
RoJo


My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
Joined: Jun 2005
Posts: 47
Technologist
Offline
Technologist
Joined: Jun 2005
Posts: 47
Hi we've got exactly the same problem but were awaiting the launch of the VP from Carefusion as our wards 'require' inline pressure monitoring, we too work on the 50mmHg default settings.

Joined: Aug 2007
Posts: 306
Master
Offline
Master
Joined: Aug 2007
Posts: 306
We have been waiting for the paed version from carefusion for nearly 4 years now so I wouldnt hold your breath.

Joined: Jan 2003
Posts: 39
Visionary
Offline
Visionary
Joined: Jan 2003
Posts: 39
Hi David.F

We use the Fresenius Agilia here at out hospital. NICU use it as well as adult areas. The pressure management on the pumps can be set for use in the NICU. The pump can either be set up into one of two pressure management modes, namely VARIABLE MODE or 3 LEVELS MODE.

VARIABLE MODE would be used in the NICU, as you would set the pump a maximum pressure between 50 to 750 mmHg, in 50 mmHg increments. This defines the authorized maximum pressure during infusion.

Also, a pre-alarm level of 50 mmHg below maximum pressure is present. It warns the user when the pressure reaches within 50 mmHg of the maximum. If the maximum pressure is set at 50 mmHg, then the pre-alarm is set at 25 mmHg.

Also, the pumps can be set-up on MICRO mode, to allow increments of 0.1 ml for volume and rate settings.

The pumps are generally good. We don't get a lot of issues with them and they tend to be very accurate. Biggest problem seems to be general deterioration due to heavy use.


Last edited by Biomed; 23/02/11 5:33 PM.

The chief function of your body is to carry your brain around.
Joined: Aug 2009
Posts: 41
Technologist
Offline
Technologist
Joined: Aug 2009
Posts: 41
Afternoon Dave,

We just recently changed to frasanius. Any questions? just drop me an email

Dee

Joined: Feb 2011
Posts: 3
David.F Offline OP
Newbie
OP Offline
Newbie
Joined: Feb 2011
Posts: 3
Thanks all,
I have managed to convince the Paediatric dept to go for a 100mmHg starting point because of the prealarm and the speed of response, so we'll see how it goes. 4 week trial starting with training next week.
Thanks for all of your replies.
Cheerio, Dave.

Joined: Jun 2001
Posts: 246
Master
Offline
Master
Joined: Jun 2001
Posts: 246
Hi,
Just in the process of finalising this and would love to se the results from there and can you send them over. We are havinga big headache with tenders for volumetrics and have come up against 'very bad trial,sales team and after sales even before the purchase so please fel free to email me. paul.kirkby@kingstonhospital.nhs.uk
Thanks
PAul


ERRATIC MEANS STATIC SO BE ERRATIC AND NOT STATIC
WE ARE ALL IN THE NHS AND THIS IS
"ERRATIC AND STATIC"

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
3 members (daisizhou, Neil Porter, Huw), 10,991 guests, and 17 robots.
Key: Admin, Global Mod, Mod
Newest Members
AndySkippy, happyvet, Abdulaziz abed, Fernando Henz, Nick.A
10,375 Registered Users
Forum Statistics
Forums26
Topics11,264
Posts74,540
Members10,375
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5