Hi Brian,
We currently have 80 single channel Signitures and one recently purchased dual channel. Overall we have found these pumps to be extremely reliable and have not found any major short comings with them.
The event log facility we have found to be invaluable in proving the sequence of events in clinical incidents and also trying to understand what staff have been doing when you get that obscure fault reported that usually results in the scratching of heads !
Battery management has sometimes been an issue but if you are aware that the pump has its own battery refresh facility (NiCad cycle) and is performed when PPM'd there is no problem. Our own pumps of which the oldest are two and half years old are just starting to require battery replacement which I don't think is bad for a NiCad battery.
Calibration of the pump is both quick and uncomplicated and does not require any special equipment other than the obligatory IDA/Burette and digital monometer. Also worth mentioning is the manufacturers recomended PPM interval is 12 months.
If you want to contact myself or John who you know to have a more in depth chat please don't hesitate to give us a ring Brian.