|
|
Joined: Apr 2010
Posts: 33
Visionary
|
OP
Visionary
Joined: Apr 2010
Posts: 33 |
Hello friends,
We have a couple of Propaq monitors with IBP and Etco2 Function, can anyone please help me on how to check on them within our biomedical department, prior being used on patient into the OT !
Best Regards Nimish nimishpancholi@yahoo.com
|
|
|
|
|
Joined: Feb 2004
Posts: 14,801 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,801 Likes: 72 |
IBP can be checked out with an appropriate tester. EtCO2 is checked with Cal Gas. 
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Feb 2003
Posts: 362 Likes: 3
Sage
|
Sage
Joined: Feb 2003
Posts: 362 Likes: 3 |
Ideally you check the functions using the manufacturers recommended procedures. In the absence of this and with limited equipment the following can be used for verification purposes.
IBP leads usually consist of 4 connections, a positive and a negative pressure signal input, and to the monitor (+/-) a positive and a negative excitation voltage output from the Propaq monitor. The IBP module supplies a differential excitation voltage to the pressure transducer that connects to the IBP lead. The monitor provides this excitation voltage to the pressure transducer
The pressure transducer usually gives a differential output voltage from the bridge of 5uV/V/mmHg or 40uV/V/mmhg i.e. 5/40uV per volt excitation per mmHg pressure applied to the transducer.
Once you know the IBP pin outs, it is relatively easy to simulate static pressures in older monitor types. (Newer monitors will not give you a reading unless it senses an attached transducer. Build a bridge with 1k resistors in the legs and an extra 100R in the fourth leg with a short out switch across it).
So check of IBP consists of a Cal mode static pressure zero mmHg and a span (say 300 mmHg)
So with 5uV sensitivity, and an excitation voltage of 10.000 volt, (you use a dc calibrator, or dc power supplies and a multimeter with suitable resolution) you can provide and provide a simulated pressure signal of 1.5 mv across the positive and negative pressure signal inputs this should simulate a static pressure of 300 mmHg.
0.000005 X 10 X 300 = 0.0015 Volts (1.5 mV)
Like wise with a 40uV sensitivity
0.000040 X 10 X 300 = 0.0120 Volts (12.0 mV)
Alternatively if you have an IBP transducer, check zero and span with a manometer.
EtCo2, ideally you need a sample line, calibration gas (nominal 5% CO2) and a regulator and follow the cal procedure. A quick verification can be performed by sampling exhaled breath after a few exhalations into the sample line. Check the monitor EtCo2 waveform is detecting Co2 and check numeric readings.
Many years ago, when we didn’t have all the flash test equipment we had to make do with what was available.
Happy Days....
|
|
|
|
|
Joined: Feb 2004
Posts: 14,801 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,801 Likes: 72 |
Good stuff, Sean.  However, "many years ago" we didn't have EtCO2 either, as far as I recall!
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Jun 2009
Posts: 796 Likes: 13
Philosopher
|
Philosopher
Joined: Jun 2009
Posts: 796 Likes: 13 |
Base principles, what a wonderful thing.
IBP, can be tested with nothing more than a cable and transducer from the system, a three-way tap, some tube, a manual bulb and a calibrated pressure meter.
Connect all three together, pump up the bulb until the pressure meter show (say) 100mmHg and check that the monitor says the same. A static measurement I will grant you but a "down and dirty" way of doing checks. In modern systems, if this is accurate, so will the dynamic readings be.
As for ETCo2, cal gas is only good if we are talking side-stream however, mainstream systems often have a "cal check" window system attached. If nout else is available, I would go with the "breathe down the sensor and see what results you get" method. Not auditor friendly but very accurate (check with the nursing staff for correct results 'cos I can't remember).
Ok Geoff, ........... over to you for why this is all wrong
Thoughts and information provided on this forum are mine and mine alone and do not necessarily reflect the policy of NSW Health. They may also be complete bollocks!!
|
|
|
|
|
Joined: Feb 2004
Posts: 14,801 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,801 Likes: 72 |
If nout else is available ... No problems* here, Mate. That's the default situation for many of us, after all. In modern systems, if this is accurate, so will the dynamic readings be. Yes. Very true. Another way of testing IBP is if you happen to have a spare artery hanging around somewhere!  * Apart from spelling, of course (nowt)!
If you don't inspect ... don't expect.
|
|
|
|
|
Joined: Apr 2010
Posts: 33
Visionary
|
OP
Visionary
Joined: Apr 2010
Posts: 33 |
Dear Friends
I was busy in some other work so could not check on it, guys thanks to all of you for describing this procedure we will start tom. morning on this and will let you know all what happened next...
sean, geoff, davec .. many thanks again..
Best Regards Nimish
|
|
|
|
|
Joined: Feb 2001
Posts: 35
Visionary
|
Visionary
Joined: Feb 2001
Posts: 35 |
Hi Nimish You can download the service manual from the Welch Allyn web site; Propaq Encore Propaq CS Most of our manuals are available this way, just find the product. And go to the "Technical" tab on the box at the bottom of the page.
Jonathan Pope Medical Systems Engineer Welch Allyn UK
|
|
|
|
|
Joined: Jul 2002
Posts: 2,020
Hero
|
Hero
Joined: Jul 2002
Posts: 2,020 |
I wish more manufacturers did this - and the same for user manuals. Thank you WA RoJo
My spelling is not bad. I am typing this on a Medigenic keyboard and I blame that for all my typos.
|
|
|
|
|
Joined: Feb 2004
Posts: 14,801 Likes: 72
Super Hero
|
Super Hero
Joined: Feb 2004
Posts: 14,801 Likes: 72 |
If you don't inspect ... don't expect.
|
|
|
|
Forums26
Topics11,253
Posts74,502
Members10,363
| |
Most Online59,530 Apr 30th, 2026
|
|
|
|