Having taken a closer look, I have been able to get this one sorted, and CAL'd the test lead to be able to carry on with further testing.

I found that the real culprit was a "loose" spade connection at the rear of the red jack socket.

But did you mean 0.06 volts AC, Sean ... as that's all I seem to be getting across the red and green jacks. Test lead resistance now comes up as 46 milliohms, by the way. smile


If you don't inspect ... don't expect.
Joined: Feb 2003
Posts: 362
Likes: 3
Sage
Offline
Sage
Joined: Feb 2003
Posts: 362
Likes: 3
Glad you got it sorted!
Re-checked the expected voltages during cal with our old and trusty ISA601PROUKPK (the sloped front one with the side cartridge).
As expected, in both 1 Amp and 10 Amp Cal mode, 6.0 Volts AC.
Is yours a different model?
Soldering a loose spade connector to the socket is an acceptable solution.
(Has been a known problem with these beauties).

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Mornin' Sean smile

This one is a ISA601PROCUKP ... sloped top, yes. No side cartridge, though. Software version 1.08. Maybe I misread the voltage. Happily, it seems to be working as expected now.

There is another one lying around here that is a ISA601PROCUSP (with 3.00 software). But that is the one I may have mentioned before. It can't seem to make up its mind about mains voltage (it starts off displaying around 150 volts at L1 to earth (and L1 to L2), then proceeds to "count up" to 230). It also definitely fails any attempt at a test lead cal. As in:- "CAL FAILED" (can't be much clearer than that).

This second one has another nice feature:- it doesn't always power up nicely, but sometimes prefers to cough (or, if you like, buzz, rattle or chunter), issue smoke, and smell a bit! whistle

But when it does fire up OK, it generally comes up with something like:- "Self-test: Warning - mains voltage out of range"! Then, after pressing [Esc] and then [Voltage] (test 1), off it goes, "counting up the mains voltage", as mentioned above. It gets there after a minute or three.

I suspect that these beasts, like many other bits of kit I have encountered over the years (and especially stuff with relays, transformers etc. inside) benefit from regular exercise, and not just lolling around for months (years?) without being used.


If you don't inspect ... don't expect.
Joined: Apr 2009
Posts: 81
Adept
Offline
Adept
Joined: Apr 2009
Posts: 81
my department has buy this machine too.
I sometime see a same problem and disply that.
"Warning: Open Earth on Mains".

Last edited by dennis_yim; 10/03/11 8:40 AM.
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Offline
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

All the clues are there, Dennis! smile

I expect that these beasts can be put right, but you need to set aside an hour or two with nothing else to do but "dig" through them. Remaking connections, cleaning contacts ... and stuff like that.

And then, once you have them working, (as I say) I reckon they need regular exercise!


If you don't inspect ... don't expect.
Joined: Jul 2010
Posts: 84
Adept
Offline
Adept
Joined: Jul 2010
Posts: 84
My office has got 2 of the same recently.. So i'd better get started on these right away.

Geoff sir,
can you make a flow chart, of sorts, to rectify this problem so that it can be readily used as and when required... Just a humble request..

Thanks.

Joined: Jul 2010
Posts: 84
Adept
Offline
Adept
Joined: Jul 2010
Posts: 84
Well someone here had requested the link for the service manual of the beastly 601pro.. Please find it below..
Courtesy: The Reza Najafi Website.
The 601 pro XL Service Manual

Joined: Feb 2003
Posts: 362
Likes: 3
Sage
Offline
Sage
Joined: Feb 2003
Posts: 362
Likes: 3
Looks like the Operators manual to me.
Your hyperlink is misleading
The 601PRO service manuals are not in the public domain.

Sometimes, you just have to get on with a repair job yourself.
If you don’t have the knowledge or experience, procure the service from the appointed agents. A copy of a service manual without specialist knowledge, calibration jigs and test equipment won’t get you very far.

In this game you need to know what you can safely do yourself, and recognise what work you must leave to the experts.


Joined: Jul 2010
Posts: 84
Adept
Offline
Adept
Joined: Jul 2010
Posts: 84
OOPS...
Sean sir, My bad.. its a typo.. intended to say operator but came out as service...

Can the moderators kindly edit it accordingly..

Joined: Feb 2007
Posts: 303
Likes: 17
Master
Offline
Master
Joined: Feb 2007
Posts: 303
Likes: 17
Originally Posted By: Geoff Hannis

But did you mean 0.06 volts AC, Sean ... as that's all I seem to be getting across the red and green jacks. Test lead resistance now comes up as 46 milliohms, by the way. smile


If you only get 0.06V I would measure the test current supplied by the meter as I have seen other units with this fault and although the unit displayed the test current as 1.04A it actually measured 0.56A! so can lead to erroneous readings.

Maybe this is a common fault on units that are quite old and well used?

Also if you obtain a resistance reading that fails the device under test may in fact be perfectly fine but contain an inductor in the earth connection. Many meters, including this one, that use an AC test current, can give false readings when an inductor is present in the earth!

Page 3 of 4 1 2 3 4

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (vitapablo), 5,096 guests, and 162 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,290
Posts74,597
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5