Did the device bring up an error code in the error log after exhibiting this fault? we have never had a fault with the opto joule switch...yet, but in the past have had problems with the pacer keypad not working and passing the daily check. This didnt leave an error code in the error log either. Given the limitations of the current self-test sequence on this device, our concerns are that it is not made sufficiently clear to the users that the this self-test does not verify functionality completely. After all the member of clinical staff that undertakes the daily shift check would consider the defib ready for use in an emergency if the test strip tells them it has passed all tests, when in fact it is faulty. We have expressed our concerns to Philips with regards to this device not picking up internal faults and had an unsatisfatory response..(again). My advice is to report this matter to the MHRA as Mr Ling says and let them add it to their surveillance database.