Home Articles Downloads Forum Products Services EBME Expo Contact ☰
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 6 of 8 1 2 3 4 5 6 7 8
Joined: Jul 2009
Posts: 79
Adept
Offline
Adept
Joined: Jul 2009
Posts: 79
IntelliVue XDS Application revision H.15 (Build 342) is now available for download from the XDS website: http://intellivue.medical.philips.com/download

This release adds important features for XDS usage on MX800/iPC combination, as well as XDS running on Philips 865159 PCs with other supported monitors.

Features for MX800/iPC:
1. H15 release adds iPC desktop mode for the IntelliVue MX800 patient monitor. When you touch a Smart Key to launch an application, the patient monitor display automatically compacts in to a "real time column", giving the PC up to 1400 x 1040 pixel desktop space, allowing you to operate any 3rd party charting system, PACS system or other application that requires a large viewing area.
The attached image "iPC desktop-1.png" shows how this looks like.
2. Input device sharing has been enhanced. The mouse connected to the iPC can also operate the MX800 monitor, and the cursor moves from the transparency mode window or iPC desktop seamlessly in to the patient monitor area. A click anywhere on the screen will be sent to the monitor or the iPC as appropriate. Keyboard input will be directed to the device which currently "owns" the mouse.

Features for XDS running on Philips 865159 XDS Bedside PC:
1. Auto-connect to any monitor connected to a specific LAN2 port (LAN2A or LAN2B) - for example, if you bring a patient on an IntelliVue X2 monitor to a bed, simply connect the monitor to the designated LAN2 port, and the monitor will auto-connect.
2.When 2 displays are connected to Philips 865159 XDS Bedside PC, each of the LAN2 ports can be assigned to a display - eg. a monitor connected to LAN2A port can be assigned to always appear on the left display, while the one connected to LAN2B will always appear on the right display. See attached diagram "port association.png"

If you have any questions, please don't hesitate to ask.

Attached Images
iPC desktop-1.png (420.69 KB, 4424 downloads)
port association.png (110.22 KB, 5047 downloads)

Chandana Samaranayake MD

Chandana Samaranayake is an employee of Philips Healthcare.
All comments made on this forum are made on a personal capacity, and do not necessarily represent the views of my employer.


As per FDA guidelines on on-line communication, please note that the Intended Use of a Philips IntelliVue product is as stated in the Instructions for Use document.
Joined: Oct 2006
Posts: 93
Mike_d Offline OP
Adept
OP Offline
Adept
Joined: Oct 2006
Posts: 93
Hi Chandana, does the "real time column" only work with an MX800 or can this be configured on a standalone PC running the XDS software connected to a philips monitor.

Joined: Jul 2009
Posts: 79
Adept
Offline
Adept
Joined: Jul 2009
Posts: 79
Hi Mike, the real-time column is only meant to work with MX800/iPC at this time. However, it would be interesting to explore the idea of using this type of screen on XDS running on a standard PC connected to other monitors, such as X2.


Chandana Samaranayake MD

Chandana Samaranayake is an employee of Philips Healthcare.
All comments made on this forum are made on a personal capacity, and do not necessarily represent the views of my employer.


As per FDA guidelines on on-line communication, please note that the Intended Use of a Philips IntelliVue product is as stated in the Instructions for Use document.
Joined: Jul 2011
Posts: 16
Novice
Offline
Novice
Joined: Jul 2011
Posts: 16
Hi, if i using XDS PC in bed 1 connected to intellivue X2 to enlarge screen of X2, can i also view the intellivue X2 information on the Intellivue center at nurse station. thank you for help

Joined: Jul 2009
Posts: 79
Adept
Offline
Adept
Joined: Jul 2009
Posts: 79
Hi HC1528,

Yes you can. In this case, you connect the X2 via LAN in to the IntelliVue Clinical Network, and you also connect the PC via LAN to the same network. This set up is described as "Topology 2: standalone subnet" in the XDS Installation and Config guide.

In the XDS configuration, you need to select the checkbox that says "IntelliVue Information Center on the network". This will disable any XDS services that conflict with the central station.

The X2 will be viewable at the central station, as well as on the XDS. If you configure the XDS application to auto-connect to the X2 in such a set up, I recommend using Bed Label as the identifier for the default connection. This way, no matter which X2 comes to the bedside, as soon as the central station gives the monitor the bed label for the location, the XDS application will auto-connect to it.


Chandana Samaranayake MD

Chandana Samaranayake is an employee of Philips Healthcare.
All comments made on this forum are made on a personal capacity, and do not necessarily represent the views of my employer.


As per FDA guidelines on on-line communication, please note that the Intended Use of a Philips IntelliVue product is as stated in the Instructions for Use document.
Joined: Jul 2011
Posts: 16
Novice
Offline
Novice
Joined: Jul 2011
Posts: 16
Hi Chandana,
thanks for the respone.
for this statement "I recommend using Bed Label as the identifier for the default connection" you are mention.

Q1, is this needs to configure the central station paired with network port number and Bed Lable.

Q2, if i have HL7 output at central station and central connected to ICN network. XDS box at Bed1 also connected to ICN network. then i have few X2 (different equipment lable) connect to Bed1's XDS in different timing, how about Bed Label and equipment lable in the HL7 message sending out from central station.
thanks for help



Joined: Jul 2011
Posts: 16
Novice
Offline
Novice
Joined: Jul 2011
Posts: 16
Q3, can i use XDS software H.15 (build 342) remote connect to MP70 with Software Rev H.03 and X2 with Software Rev G.01.75. will it cause any conflit (software) problem. any suggestion from you

Joined: Jul 2009
Posts: 79
Adept
Offline
Adept
Joined: Jul 2009
Posts: 79

Q1, is this needs to configure the central station paired with network port number and Bed Lable?

A: Yes. The bed label needs to be configured for the patient monitor's network port number at the central station. This means, every time you connect any monitor to this port, it will receive the same bed label. Since you have configured the XDS to auto-connect to any monitor with this bed label - this results in XDS auto connecting to any monitor connected to this port.

Q2, if i have HL7 output at central station and central connected to ICN network. XDS box at Bed1 also connected to ICN network. then i have few X2 (different equipment label) connect to Bed1's XDS in different timing, how about Bed Label and equipment label in the HL7 message sending out from central station.

A: In this case, any X2 being connected to this network port will get the bed label Bed1. XDS will auto-connect to any of these monitors. The central will send out the data from the monitor via HL7 with the bed label bed 1, but also with the equipment label configured on the monitor, eg: Monitor with equipment label mon7 will get bed label Bed1, the HL7 message will contain "|Bed 1|Mon7|". This represents the true origin of the data - from bed 1, using the equipment Mon7.


Chandana Samaranayake MD

Chandana Samaranayake is an employee of Philips Healthcare.
All comments made on this forum are made on a personal capacity, and do not necessarily represent the views of my employer.


As per FDA guidelines on on-line communication, please note that the Intended Use of a Philips IntelliVue product is as stated in the Instructions for Use document.
Joined: Jul 2009
Posts: 79
Adept
Offline
Adept
Joined: Jul 2009
Posts: 79
Q3, can i use XDS software H.15 (build 342) remote connect to MP70 with Software Rev H.03 and X2 with Software Rev G.01.75. will it cause any conflit (software) problem. any suggestion from you

A: This will not cause a conflict. When X2 is docked to an MP70, it transforms in to a measurement acquisition module for the host monitor (this is called "companion mode"). With the X2 in companion mode, it will not even be visible to the XDS. XDS will only "see" the MP70. X2 with rev G.xx is fully compatible with an H.xx host monitor.

That said, my recommendation would be for you to upgrade the X2 to H.xx. It is always best to keep the X2s up to date, as we continually improve our measurement algorithms with each release. As you may be aware, software upgrades for X2 are free of charge, and can be done via the IntelliVue Support Tool using your Biomed license key. If you request Philips to do the upgrade for you, you will be billed for time and travel, but no software upgrade needs to be purchased.


Chandana Samaranayake MD

Chandana Samaranayake is an employee of Philips Healthcare.
All comments made on this forum are made on a personal capacity, and do not necessarily represent the views of my employer.


As per FDA guidelines on on-line communication, please note that the Intended Use of a Philips IntelliVue product is as stated in the Instructions for Use document.
Joined: Jul 2011
Posts: 16
Novice
Offline
Novice
Joined: Jul 2011
Posts: 16
Hi Chandana Samaranayake,
currently, i using XDS software (Build 303) (installed on HP branded PC) to remote view/control the MP70 in the ICU area. From you previous post, Philips has new release of XDS software (Build 342), should i upgrade the current software version to latest and any different after install new software.

if IT department has regular upgrade windows XP operating patches / .NET framework / inhouse software, should i also upgrade the XDS software when have new release.
thanks for help

Page 6 of 8 1 2 3 4 5 6 7 8

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (daisizhou), 13,315 guests, and 291 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,287
Posts74,592
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5