Home Articles Downloads Forum Products Services EBME Expo Contact
Previous Thread
Next Thread
Print Thread
Rate Thread
#57480 02/08/11 3:16 PM
Joined: Jul 2002
Posts: 29
Dreamer
OP Offline
Dreamer
Joined: Jul 2002
Posts: 29
Hi All
We are currently reviewing the frequency of our Electrical Safety testing. What frequencies do other Trusts carry this out at?

Thanks
Mandy

Joined: Jul 2011
Posts: 66
CHJ Offline
Scholar
Offline
Scholar
Joined: Jul 2011
Posts: 66
Hi Mandy,

Is this in relation to just Medical Electrical Safety (IEC 62353) or the wider view of Portable Appliance Testing in general?

I'd be happy to share my knowledge of what I know from Electrical Regulations (having spent 3 years selling Electrical Safety before working for JPen)

Kindest regards,


Chris Horwood-Jones
Joined: Jul 2002
Posts: 29
Dreamer
OP Offline
Dreamer
Joined: Jul 2002
Posts: 29
Hi Chris
This would be in the wider aspect, if you could email me I would be grateful?

Many thanks

Mandy

Joined: Jul 2011
Posts: 66
CHJ Offline
Scholar
Offline
Scholar
Joined: Jul 2011
Posts: 66
Hi Mandy,

Check your email.

Hope it helps.

Kindest regards,


Chris Horwood-Jones
Joined: Jun 2000
Posts: 2,412
Likes: 12
Huw Offline
Hero
Offline
Hero
Joined: Jun 2000
Posts: 2,412
Likes: 12
Hi Jonesy,
I think it would be helpful to the wider readership of this forum if you could make that email public. smile

Joined: Jul 2011
Posts: 66
CHJ Offline
Scholar
Offline
Scholar
Joined: Jul 2011
Posts: 66
Hi Huw - my pleasure smile

Dear all - for your reference please see below.

Portable Appliance Testing, or PAT testing as it is commonly referred to, has always been a bit of a grey area.

There has never been an official legal requirement to do PAT testing. The main need for having to do PAT testing is under the Health and Safety at Work Act 1974 (HASAWA) where employers have a “duty of care” to their staff and visitors.

The Electricity at Work Regulations (EWR) state “As may be necessary to prevent danger, all systems shall be maintained so as to prevent, so far as is reasonably practicable, such danger.”

So far as frequency is concerned, this can only be determined by the Duty Holder. Many companies view best practice to be testing every appliance every year, though this needn’t be the case if you manage the appliances and keep track of them.

The HSE did release some guidance on the frequency of PAT testing, as did the IEE (Institute of Electrical Engineers). However, this is only generic and is not necessarily suitable to each organisation.

The best advise I can provide is run a series of risk assessments for different types of equipment and different work areas. It is reasonable to suggest that a PC, which is rarely moved, provides a lower level of risk than that of a cordless drill that is obviously much more portable and much more likely to cause an injury through electric shock.

I recall some companies having testing spread over a 4 year period as follows:

Year 1 – test on ALL equipment
Year 2 – test on all HIGH RISK equipment
Year 3 – test on all MEDIUM and HIGH risk equipment and a visual only inspection on LOW RISK equipment
Year 4 – test on all HIGH RISK equipment

And then repeat the whole thing over again.

The most important thing is to have the risk assessments conducted so that you have a paper trail detailing the choices that have been made and why – so that it is clear the decision was based on a risk assessment and not just simply “not doing this year because I can’t afford it”.

The main thing to consider is, should the worst happen and you get taken to court for a faulty piece of equipment and they ask for PAT testing records, would you be happy to stand in court with your evidence and say that you took all “reasonably practicable” steps to reduce such danger. If you can satisfy yourself (or the Duty Holder) that the answer is yes then you should be able to rest easy.

You might want to consider speaking with NHS LA (the NHS Litigation Authority) as they may have more specific advice concerning PAT testing in the NHS. I know that they have key standards when it comes to managing medical devices but have not spoken to them about PAT testing and Electrical Safety Testing in general.

I hope this information has been of some help, but if anyone has any other questions, please do not hesitate to get in touch with me.

Kindest regards,


Chris Horwood-Jones
Joined: Jun 2007
Posts: 98
Adept
Offline
Adept
Joined: Jun 2007
Posts: 98
May I say what an excellent post Chris has made concerning electrical testing of portable appliances. My personal view is that most of it has become a sham and much of it is unnecessary whilst other more important elements are ignored. It seems mistaken to me that most hospitals will be more concerned in testing the appliance and care little about the IEC lead. In my experience it is the mains lead that is the weak point of all portable appliances and integrity of the earth continuity together with a thorough visual inspection will weed out 99% of dangerous occurances.


Barney
Joined: Dec 2003
Posts: 7
Newbie
Offline
Newbie
Joined: Dec 2003
Posts: 7
Hi,
I was just wondering with respect to this post; are most departments testing to IEC 62353 or still testing to IEC 60601?

A


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
2 members (daisizhou, KiwiSi), 3,027 guests, and 26 robots.
Key: Admin, Global Mod, Mod
Newest Members
Chris 11, j9_PLC, nece, Vitya, Shenzhen007
10,358 Registered Users
Forum Statistics
Forums26
Topics11,248
Posts74,481
Members10,357
Most Online37,242
Apr 12th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5