Home Articles Downloads Forum Products Services EBME Expo Contact ☰
Previous Thread
Next Thread
Print Thread
Rate Thread
Page 1 of 2 1 2
Joined: Jul 2005
Posts: 601
Philosopher
OP Offline
Philosopher
Joined: Jul 2005
Posts: 601
Hi,

Does anyone carry out electrical safety tests on Battery operated AED Defibs? I was told that we should be doing patient auxilliary current tests between the outputs. Looking at the safety tests in 60601 only the normal condition applies to this device (can't reverse mains or lift neutral or earth). Is it worhwhile measuring the aux. current in normal condition?

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Online Content
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

It's a defib. It's meant to provide an "electric shock" to the patient! smile


If you don't inspect ... don't expect.
Joined: Jul 2005
Posts: 601
Philosopher
OP Offline
Philosopher
Joined: Jul 2005
Posts: 601
It's also an ECG monitor!

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Online Content
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

It's also battery powered! smile

What intrigues me more here is where this is coming from. As in, who is telling you that the test you mention is needed? The "management" ... the manufacturer?

Is the requirement "evidence based" (have these things been known to "fail" on patient auxilliary current)?

Or, put another way, which way are we looking at things:-

1) If it ain't bloke, don't fix it.

... or:-

2) If it ain't bloke, you're not trying hard enough!


If you don't inspect ... don't expect.
Joined: Jul 2005
Posts: 601
Philosopher
OP Offline
Philosopher
Joined: Jul 2005
Posts: 601
One of my colleagues has just retired and I've taken on his workload (on top of my own!). He did the leakage tests, didn't really explain who told him to do them. I'm of the opinion that it is unnecessary, but just wanted to see if anyone else did it.

Joined: Dec 2010
Posts: 22
Dreamer
Offline
Dreamer
Joined: Dec 2010
Posts: 22
Originally Posted By: Geoff Hannis
Is the requirement "evidence based" (have these things been known to "fail" on patient auxilliary current)?

If you are looking for failures, then thats not preventive maint it's just corrective.

I have done EST on battery powered defibs. If you want zero harm from an equipment, you might as well do it. It's a high risk equipment; can cause big damamge if it fails.

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Online Content
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Fair enough (except that I don't really follow what you're saying at your first line). smile

Which tests do you carry out?

(we're talking about AED's here)


If you don't inspect ... don't expect.
Joined: Jul 2005
Posts: 601
Philosopher
OP Offline
Philosopher
Joined: Jul 2005
Posts: 601
The only test we can do is "Patient Auxilliary Current" under "Normal" conditions.

Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Online Content
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Yes. But let's hear more from Mr. Tejas. smile


If you don't inspect ... don't expect.
Joined: Feb 2004
Posts: 14,815
Likes: 72
Super Hero
Online Content
Super Hero
Joined: Feb 2004
Posts: 14,815
Likes: 72

Originally Posted By: biomedbill
One of my colleagues has just retired and I've taken on his workload (on top of my own!).

If you need a hand Bill, I can make myself available. smile

My rates are very reasonable. "Dross"* work, and out-of-hours working are specialities! Just find me somewhere to kip (hint:- it doesn't have to be at the back of the workshop).

And it would be nice passing by old haunts in nearby Rotherham! Many of my chav mates are there. whistle

Meanwhile, and by the way, which battery-powered AED's are we talking about? You know, manufacturer, model? I'm only asking in case I need to do a bit of research and/or if I come across one I can take a look at.

Having looked through a couple of manuals at random recently, I have yet to see any reference to the test you mention. One mentions "Leakage current checks to IEC 60601-2-4" (surely a catch-all ... and no mention about how such checks are to be carried out). Whilst another makes no mention of any periodic tests beyond the "Self Test".

* You know, the stuff that everyone else is always too busy to get around to doing.


If you don't inspect ... don't expect.
Page 1 of 2 1 2

Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
1 members (Geoff Hannis), 7,845 guests, and 298 robots.
Key: Admin, Global Mod, Mod
Newest Members
Pip Fox, Engzaki, MacMc, AraEmin, Edouardohaag
10,395 Registered Users
Forum Statistics
Forums26
Topics11,285
Posts74,589
Members10,395
Most Online69,877
Jul 17th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5