Dear all
Have any one faced similar fault.
I have replaced the faulty mosfets (APT8024 & IRF840) & fuses but still having similar error.
Error 160 means internal component fault DOSAGE ERROR.
Can some one guide which area to look for dosage error. As during bootup the machine takes more time when checking CUT option during start up.
Hope some one can help me out here.
Thanks in advance.

Laiq #59467 12/12/11 11:30 AM
Joined: Nov 2010
Posts: 164
Mentor
Offline
Mentor
Joined: Nov 2010
Posts: 164
Mosfets and fuses that you have replaced is in the Power Supply Section?...There might be other defective components away from the PSU, try to look in the Output Portion mainly on the Cut circuitry...


There is no work worth dying for.
Joined: Oct 2007
Posts: 75
Adept
Offline
Adept
Joined: Oct 2007
Posts: 75
Don’t now if this will help, as I have only worked on the force fx, dosage errors 16X I have found have been caused by breast surgery, the surgeons driving the excessively, causing the to over heat, but the have found if the switch off the unit and back on it clears the error and they carry on using it until it eventually fails. If you have replaced the fets, be careful and check there is no excessive ringing on the rf output stage when activated (I think it’s the drain of the FET TP19) this should be done once the fet is replaced at low power, if done at high power you will blow the FET again. If there is ringing the output transformers and the capacitors should be replaced. Note this is based on the force fx I don’t know how comparable they are.

Joined: Aug 2010
Posts: 5
Newbie
Offline
Newbie
Joined: Aug 2010
Posts: 5
I have fixed a couple of units with error 160 by replacing the control board but they are ForceFX. I am thinking it may be similar!

Last edited by Keeglor; 30/12/11 7:45 AM.
Joined: Apr 2005
Posts: 380
Laiq Offline OP
Sage
OP Offline
Sage
Joined: Apr 2005
Posts: 380
Thanks guys. That particular ESU was being used in GYNEA OT.
Waiting for replacement control board. :-)


Moderated by  DaveC in Oz, RoJo 

Link Copied to Clipboard
Who's Online Now
2 members (vitapablo, Neil Porter), 1,205 guests, and 27 robots.
Key: Admin, Global Mod, Mod
Newest Members
Nick.A, Tdayve, AdrianT82, Iagonuns, Dctm
10,371 Registered Users
Forum Statistics
Forums26
Topics11,261
Posts74,525
Members10,371
Most Online59,530
Apr 30th, 2026
Powered by UBB.threads™ PHP Forum Software 7.7.5