i have a problem with GE DASH 4000 Patient Monitor.Its showing an error message "SERVICE MONITOR, Possible Data Corruption" . Does anyone know about it??
Hi, if you monitor is connected to network, may be you have communication problems. I not connected, so your monitor have activated network option. You need acces service mode and clear errors. If you need service manual, send me your email to hugo_68@yahoo.com
hi, the monitor is not connected to any network,and the wireless LAN is disabled. Cleared the software errors from service menu. Still the problem persists..!! any other solutions?
We have had this happen a few times, it's been a while but try Doing a "COLD START" (Press & hold "Graph Go/Stop button while switching the Monitor on, hold button down for a couple of seconds, the Monitor screen will flash then come on) Alarms may reset to default and Patient data will be lost, so you will need to check / reset Alarm settings.
Hope this will help, Regards,
Mark